Tektronix Technical Forums are maintained by community involvement. Feel free to post questions or respond to questions by other members. Should you require a time-sensitive answer, please contact your local Tektronix support center here.

Search found 196 matches

by RaymondR
September 25th, 2013, 11:07 am
Forum: 4200-SCS Semiconductor Parameter Analyzer
Topic: KXCI is not responsive when it exited on compliance reach
Replies: 4
Views: 18858

Re: KXCI is not responsive when it exited on compliance reac

Hi maheshkakuturu,

Can you send me the code you're using in which this problem happens. I will try to recreate the problem on my end.
by RaymondR
September 25th, 2013, 10:31 am
Forum: Low Current Measurements
Topic: How to configure Keithley 2425 to sink current
Replies: 2
Views: 14429

Re: How to configure Keithley 2425 to sink current

Hi wcabaluna,

The "Remote Sweep Operation" section in chapter 10 of the 2400 User's Manual contains many programming examples on setting up sweeps. This section begins on page 10-19, or page 190 of the pdf.

The User's Manual can be found here: http://www.keithley.com/support/data?asset=56890.
by RaymondR
September 25th, 2013, 10:27 am
Forum: 2400 Series SourceMeter
Topic: pulse sweep 2430
Replies: 3
Views: 16352

Re: pulse sweep 2430

Hi Banjopicks, Try replacing the ":READ?" string you send with an ":INIT". Add a small delay and then send a "TRAC:DATA?" command. Also, add the following commands before sending the ":INIT" :TRAC:FEED:CONT NEV :TRAC:CLE :TRAC:POIN count :TRAC:FEED:CONT NEXT These commands will set up your data buff...
by RaymondR
August 30th, 2013, 7:10 am
Forum: Series 2000 and Series 2100 DMM's
Topic: Unable to update keithley 2100 Firmware
Replies: 2
Views: 13317

Re: Unable to update keithley 2100 Firmware

Hi Yi,

I have a no-check version for the 2100 firmware install. Please send me a PM with your contact info so I can send this to you for you to try.
by RaymondR
August 16th, 2013, 10:56 am
Forum: Series 2700 Multimeter/Data Aquisition/Switch System
Topic: Keithley MN 2701 High Speed Data
Replies: 1
Views: 10405

Re: Keithley MN 2701 High Speed Data

Hi jeremy.legasse, Which switch card are you using? You want to change the NPLC (rate option in ExceLINX) to .01 or less; disable autozero and autoranging; set the reading count to something large, like 2000 or so to start with. You will play with the delay setting to control the data rate if you're...
by RaymondR
August 16th, 2013, 10:47 am
Forum: Electrometers
Topic: 6517A as an AC Voltage Source - Possible?
Replies: 4
Views: 19136

Re: 6517A as an AC Voltage Source - Possible?

Hi marrinerm, The 2400 is capable of sourcing through approximately 2000 points per second. You can figure that 20 points per cycle is adequate to approximate a sine wave so you can get about 100 Hz sinusoid from 2400. The 2600 is capable of sweeping through approximately 20000 points per second. Th...
by RaymondR
August 16th, 2013, 7:10 am
Forum: Series 3700 Switch System/Multimeter and Plug-In Cards
Topic: 3732 Dual 4x56 configuration
Replies: 3
Views: 17421

Re: 3732 Dual 4x56 configuration

This problem is in the process of being worked on external to the forum.
by RaymondR
August 16th, 2013, 7:01 am
Forum: 2400 Series SourceMeter
Topic: Stacking Two 2400 SMU in Series
Replies: 1
Views: 10468

Re: Stacking Two 2400 SMU in Series

Hi csoon1, Yes, the safest way is to connect the Lo's of the instruments together and ground that terminal. You then source max voltage on both instruments, but opposite polarity. The difference between the Hi's of the two SMU's will be 1300 V. This is what's called a back-to-back series configurati...
by RaymondR
August 15th, 2013, 9:20 am
Forum: Current Sources
Topic: 6221/2182A delta mode -- repeating filter
Replies: 2
Views: 13434

Re: 6221/2182A delta mode -- repeating filter

Hi magnet,

Have you tried using a large digital filter (100 counts, repeating) with an NPLC setting greater at 1 or greater?

Make sure to check the "Enable Digital Filter" box or feed a true value boolean into the "Set 2182A Measurement Parameters" VI.
by RaymondR
August 15th, 2013, 7:16 am
Forum: Electrometers
Topic: 6517A as an AC Voltage Source - Possible?
Replies: 4
Views: 19136

Re: 6517A as an AC Voltage Source - Possible?

Hi marrinerm,

The 6517A is not capable of what you require. You could try continuously changing the output in steps to approximate a sine-wave via a program run on a PC but you would be limited to 1 or 2 Hz at the very most.
by RaymondR
August 15th, 2013, 7:09 am
Forum: Keithley 3390
Topic: Model 3390 pulse phase adjustment
Replies: 1
Views: 13887

Re: Model 3390 pulse phase adjustment

Hi ClaudiuStan, The pulse-mode phase is hardware limited. As for a potential solution, you may be able to use a sine-wave into a comparator (op-amp with no feedback) and adjust the amplitude to affect the duty cycle of the comparator output. The sine wave phase can be adjusted compared to a referenc...
by RaymondR
August 15th, 2013, 6:51 am
Forum: Nanotechnology Measurements
Topic: Measuring conductivity of electrospun polymer nanofibers
Replies: 2
Views: 17366

Re: Measuring conductivity of electrospun polymer nanofibers

Hi amoayeri,

I'm not aware of any supplier of custom test fixtures. You may have to build your own. The important thing is to make sure the four probes are equidistant from one to the next.
by RaymondR
August 15th, 2013, 6:44 am
Forum: 2657A High Voltage SourceMeter
Topic: Checking correct operation of SMU
Replies: 1
Views: 11668

Re: Checking correct operation of SMU

Hi TRF817, The reference manual for the 2657A contains performance verification procedures. Please refer to Appendix B-1 of the 2657A SMU Reference manual: https://www.keithley.com/support/data?asset=56976 . You will have to register at www.Keithley.com if you haven't done so already. It is free to ...
by RaymondR
August 15th, 2013, 6:40 am
Forum: Other Source Measurement Units (SMU)
Topic: Checking Keithley 237 before running tests
Replies: 1
Views: 10877

Re: Checking Keithley 237 before running tests

Hi TRF817,

The service manual for the 237 contains performance verification procedures. Please refer to section 1 of the 236/237 SMU Service Manual:

https://www.keithley.com/support/data?asset=876.

You will have to register at www.Keithley.com if you haven't done so already. It is free to do so.
by RaymondR
August 8th, 2013, 11:53 am
Forum: 2600 Series SourceMeter
Topic: Save or Export SMU Data to CSV or .txt, TSP Express
Replies: 1
Views: 10723

Re: Save or Export SMU Data to CSV or .txt, TSP Express

Hi Patrick G, Are you trying to save to a file on a PC from the instrument? The 2600 cannot access the file structure of a host PC. You can save data to a USB drive that's plugged into the 2600 using the file i/o commands: io.close() io.flush() io.input() io.open() io.output() io.read() io.type() io...
by RaymondR
August 8th, 2013, 11:42 am
Forum: Keithley DAQ Data Aquisition Products
Topic: Keithley 2602 for cyclic voltammetry
Replies: 4
Views: 21100

Re: Keithley 2602 for cyclic voltammetry

Hi Filooo, Are you using a 2602, 2602A, or a 2602B? We do have LabVIEW drivers for all three instruments. You will have to set up a voltage list sweep and input the voltage points comprising your waveform. The number of points depends on how closely you want the output to linearly ramp. The timing w...
by RaymondR
June 12th, 2013, 8:45 am
Forum: Current Sources
Topic: 6220 + 2182A Triggering additional DC sweeps remotely
Replies: 7
Views: 29484

Re: 6220 + 2182A Triggering additional DC sweeps remotely

Hi radadia, The PyVISA provides the python functions necessary to interface with the VISA protocol. If you're interested in using Python then I suggest taking a look at PyVISA if you haven't done so already. If you're already savvy with Python then using PyVISA will not be a problem for you. http://...
by RaymondR
June 7th, 2013, 11:36 am
Forum: IVI Drivers
Topic: ke37xx_init error
Replies: 2
Views: 17928

Re: ke37xx_init error

Hi jbodey,

What language are you writing your program in? Also, do you have the NI-VISA installed on your computer?
by RaymondR
May 14th, 2013, 12:36 pm
Forum: 2600 Series SourceMeter
Topic: Contact Check and Pulse mode Sweep
Replies: 2
Views: 13127

Re: Contact Check and Pulse mode Sweep

Hi bmiller, How long do you have when you turn off the output before starting another sweep? You can use contact check with the output off but there are some conditions that need to be met: 1. The output off mode is NOT High-Z. 2. If the output off mode is normal and the output off function is DCV t...
by RaymondR
May 9th, 2013, 12:57 pm
Forum: Current Sources
Topic: 6221 ethernet fails
Replies: 4
Views: 19452

Re: 6221 ethernet fails

Hi grochford,

The issue is currently under investigation. I was able to replicate the hang-up on firmware D01. Which firmware does your problematic instrument have?