Tektronix Technical Forums are maintained by community involvement. Feel free to post questions or respond to questions by other members. Should you require a time-sensitive answer, please contact your local Tektronix support center here.

Search found 14 matches

by picburner
November 18th, 2015, 10:21 am
Forum: Series 2300
Topic: 2303 Trip Relay mode
Replies: 2
Views: 14477

Re: 2303 Trip Relay mode

Many thanks for your prompt response.
I make the upgrade now and try the new features.
Regards.
by picburner
November 16th, 2015, 6:05 am
Forum: Series 2300
Topic: 2303 Trip Relay mode
Replies: 2
Views: 14477

2303 Trip Relay mode

I need to use the "Trip Relay mode" function but I found that in my 2303 it doesn't work because I have the firmware version A04. Where can I find the firmware A06 mentioned in the manual that enable this function? In the section dedicated to the model 2303 of your site there is only the Flashwizard...
by picburner
October 17th, 2014, 4:48 am
Forum: Logic Analyzers
Topic: Available Tek 3001 GPX full backup floppy image set
Replies: 3
Views: 13505

Re: Available Tek 3001 GPX full backup floppy image set

A good place where you can store your files is this: http://www.ko4bb.com/manuals/
by picburner
October 11th, 2014, 2:57 am
Forum: Series 2000 and Series 2100 DMM's
Topic: Model 2010 firmware
Replies: 6
Views: 19594

Re: Model 2010 firmware

Now I understand why all this mysterious behavior.
I received the response from Keithley-Tektronix Germany:
update the firmware is considered a repair for the modest sum of 1180 € + shipping + VAT.
No comment.
by picburner
October 1st, 2014, 1:20 am
Forum: Series 2000 and Series 2100 DMM's
Topic: Model 2010 firmware
Replies: 6
Views: 19594

Re: Model 2010 firmware

Hello, while doing some tests on the instrument I tried the gpib command *TST? and I got '1' as a result. If the user manual is exactly this result means that the ROM is corrupt. This would explain some instrument's incomprehensible anomalies. I realized that you want to sell the latest firmware , b...
by picburner
September 19th, 2014, 6:09 am
Forum: Series 2000 and Series 2100 DMM's
Topic: Model 2010 firmware
Replies: 6
Views: 19594

Re: Model 2010 firmware

Hello thanks for your kind reply. Maybe I have not explained it well, I know that the 2010 does not have a firmware flashable connecting it to a PC, I was wondering if you could send me the firmware in file format,then I can program the two rom myself by a eprom programmer, thanks. Best regards. Gia...
by picburner
September 19th, 2014, 1:45 am
Forum: Series 2000 and Series 2100 DMM's
Topic: Model 2010 firmware
Replies: 6
Views: 19594

Model 2010 firmware

Hello, I bought a used dmm model 2010 and I have two requests to make. Can you send me the latest firmware for this instrument in electronic format? (Intel Hex, Motorola S, no problem for me) since I have many Eprom programmers and many 27C2001, the current firmware of the instrument is A09. The ins...
by picburner
September 25th, 2013, 9:56 am
Forum: Logic Analyzers
Topic: TLA 704 setup help needed
Replies: 3
Views: 14804

Re: TLA 704 setup help needed

I am in the same situation, tla 704 with hd broken. To copy files from one machine to another you might use the tla pcmcia slots by a compact flash adapter and a cf memory. With a cf of adequate size and a program like ghost you could transfer the full hd image. If you transfer the files can send th...
by picburner
September 4th, 2013, 2:00 pm
Forum: Other or Discontinued Oscilloscopes
Topic: Help with TDS 524A Corrupted DS1650 NVRAM.
Replies: 24
Views: 70569

Re: Help with TDS 524A Corrupted DS1650 NVRAM.

Ron wrote: While working on the original NVRAM issue, I also found the NVRAM memory locations for the Flag bits that set the Calibration to Initialised or Pass. TDS524A:- Voltage Ref:- Address 0x09B7 . Frequency Response:- Address 0x09BB . Pulse Trigger :- Address 0x09BF. A value of 0x23 = Initiali...
by picburner
August 29th, 2013, 9:26 am
Forum: Other or Discontinued Oscilloscopes
Topic: How to clear error log on TDS3034B...
Replies: 2
Views: 10855

Re: How to clear error log on TDS3034B...

You can try this program.
In addition to clear the error log you can print and save it.
You need a National Instruments GPIB card.

http://www.hakanh.com/dl/TDSErrLog.htm
by picburner
April 26th, 2013, 9:50 pm
Forum: Other or Discontinued Oscilloscopes
Topic: calibration connection tips adjustment software
Replies: 4
Views: 15908

Re: calibration connection tips adjustment software

Hello sschnelle,
the patch for dosemu is needed to show the GPIB-USB-HS adapter to the FAS?
Then the program, making a similar change, it could also work well with a virtual machine like VirtualBox under any operating system.
by picburner
December 17th, 2012, 8:18 am
Forum: Other or Discontinued Oscilloscopes
Topic: TDSxxx Repair Thread list
Replies: 15
Views: 62377

Re: TDSxxx Repair Thread list

O.k. now all links in the cgifool post works well.
But still do not work the first two links of the JVG post while the third, despite having the same problem, lead in the right place.
by picburner
December 15th, 2012, 6:09 am
Forum: Other or Discontinued Oscilloscopes
Topic: TDSxxx Repair Thread list
Replies: 15
Views: 62377

Re: TDSxxx Repair Thread list

Andy is right the links does not work.
The reason is that the links lead to "http://www.tek.com/forum/..." when they should be "http://www1.tek.com/forum/..."
Probably inside the domain instead the links work normally.
If you try all work with this method.
by picburner
December 1st, 2012, 8:44 am
Forum: Other or Discontinued Oscilloscopes
Topic: TDS620B CH1 DC input problem
Replies: 6
Views: 20758

Re: TDS620B CH1 DC input problem

Hello, If the measurements made in AC are correct, you may have the AC/DC relay into the attenuator assembly broken or with oxidized contacts. When you select AC DC hear the click of the relay? Try to select AC DC several times to see if you can restore the contact. Also checks if the pin 12 of U140...