Search found 93 matches

by jaime_g
March 4th, 2019, 5:05 pm
Forum: 2600 Series SourceMeter
Topic: breakdown voltage on a Transient Voltage Suppressor
Replies: 2
Views: 2318

Re: breakdown voltage on a Transient Voltage Suppressor

Hi Andrea, I think I will use the 2410 SMU to measure the breakdown voltage of the transient voltage suppressors from Mricosemi because they have the voltage range needed (up to 1000V) to clamp the breakdown voltage measurement at 350V. I see that I can use the following scripts: 2410 source code to...
by jaime_g
February 19th, 2019, 11:29 am
Forum: 2600 Series SourceMeter
Topic: breakdown voltage on a Transient Voltage Suppressor
Replies: 2
Views: 2318

breakdown voltage on a Transient Voltage Suppressor

How can I measure the breakdown voltage of a bidirectional Transient Voltage Suppressor? The test current is 1 mA and the max Vbr specified limit is 320V
by jaime_g
November 30th, 2017, 12:26 pm
Forum: Non- Windows Based Oscilloscopes
Topic: Win10 does not recognize MDO3014
Replies: 5
Views: 13584

Re: Win10 does not recognize MDO3014

David,
Both recommendations were already implemented. Yes, the USB device Port is set to "Connect to Computer". I dis-connected the USB cable on the oscilloscope side and reconnected it again. No change.
by jaime_g
November 30th, 2017, 10:45 am
Forum: Non- Windows Based Oscilloscopes
Topic: Win10 does not recognize MDO3014
Replies: 5
Views: 13584

Re: Win10 does not recognize MDO3014

David, The NI VISA 2017 drivers are installed. I have an NI GPIB to USB interface connected to a USB 2.0 hub. I am able to see all the GPIB addresses for other instruments. Installed NI Software.docx The MDO3014 USB port is connected directly to the PC. However, the PC cannot find the instrument. Wh...
by jaime_g
November 30th, 2017, 8:46 am
Forum: Non- Windows Based Oscilloscopes
Topic: Win10 does not recognize MDO3014
Replies: 5
Views: 13584

Win10 does not recognize MDO3014

I am using a DELL Precision 3620 with Windows10 version 1709, OS Build 16299.64.
Bought a brand new MDO3014 with firmware v1.26 but, I cannot use it because Windows 10 does not recognize the instrument USB.
Windows10 Fails to Detect USB.docx
(698.17 KiB) Downloaded 888 times
Help!
by jaime_g
October 27th, 2017, 10:57 am
Forum: 2600 Series SourceMeter
Topic: How to display pulsed measurement on 2612A
Replies: 1
Views: 6430

How to display pulsed measurement on 2612A

A 2612A is configured to perform a current gain measurement on an PNP transistor. Channel A sources a negative 500µs current pulse. The pulse level, starting at -600µA is increased to -850µA in 1µA steps. The interval between 1µA steps is 100ms. The source changing amplitude is displayed on the bott...
by jaime_g
August 10th, 2017, 3:22 pm
Forum: Programming Support
Topic: MDO3014 AFG Fails to output programmed pulse width
Replies: 4
Views: 13884

Re: MDO3014 AFG Fails to output programmed pulse width

Dave, I tried the command sequence you suggested. There was no change on the outcome. However, i discovered that 10% duty cycle does not function for 600E-3s or 1.25s pulse widths. Instead, using a 20% duty cycle generated the programmed pulse (600E-3s, 720E-3s, 1.25s) correctly on the first try. Th...
by jaime_g
August 10th, 2017, 9:35 am
Forum: Programming Support
Topic: MDO3014 AFG Fails to output programmed pulse width
Replies: 4
Views: 13884

Re: MDO3014 AFG Fails to output programmed pulse width

I need to make a revision. The work around for the AFG to generate 750E-3 and 600E-3 pulse widths is changing the duty cycle of the waveform from 10% to 20% NOT 50%.
by jaime_g
August 10th, 2017, 9:07 am
Forum: Programming Support
Topic: MDO3014 AFG Fails to output programmed pulse width
Replies: 4
Views: 13884

MDO3014 AFG Fails to output programmed pulse width

When programming a pulse width of 1.25s with a period of 12.5s, i.e., 10% duty cycle, the AFG outputs a 1.0s pulse. The MDO3014 scope was returned 2 weeks ago from the factory after warranty repair was done. It is using firmware version 1.26. The following commands are used: AFG:FUNC PULS; AFG:PUL:W...
by jaime_g
May 3rd, 2017, 7:00 pm
Forum: 2600 Series SourceMeter
Topic: 2612B Triggering Pulse Measurement
Replies: 1
Views: 8555

2612B Triggering Pulse Measurement

I am trying to measure the current gain (hFE) of an NPN transistor. SMU2612B CHA V+ is used to drive the base with a 1ms, 467µA pulse limited at 5V. SMU CHB V+ is used to provide 10V (limited at 10mA) at the collector terminal. CHA & CHB V- terminals are joined together at the emitter. The oscillosc...
by jaime_g
March 9th, 2017, 9:56 am
Forum: Non- Windows Based Oscilloscopes
Topic: TDS 3052C Measurement Indicators Fail to Turn On
Replies: 3
Views: 13726

Re: TDS 3052C Measurement Indicators Fail to Turn On

Dave,

Thank you for your prompt response.
Reversing the order of the commands worked.
It is interesting to note that the original order of these commands works on the MDO3012 and MDO3014 but not on the TD3052C.

Regards,

Jaime
by jaime_g
March 7th, 2017, 4:48 pm
Forum: 2600 Series SourceMeter
Topic: 2612B End of Pulse Configuration
Replies: 1
Views: 8636

2612B End of Pulse Configuration

I have been trying configure an SMU 2612B so the end of a 15ms, 10mA pulse applied to a photo diode reaches 0V at the end of the pulse width. However, the falling edge of the pulse will not reach 0V. Instead, for 600µs the pulse hangs at about 0.2V then drops to -0.1V for another 600µs before reachi...
by jaime_g
March 7th, 2017, 4:01 pm
Forum: Non- Windows Based Oscilloscopes
Topic: TDS 3052C Measurement Indicators Fail to Turn On
Replies: 3
Views: 13726

TDS 3052C Measurement Indicators Fail to Turn On

Dear Forum, I have been trying unsuccessfully to programmatically turn on the measurement indicators for a delay measurement. Here is the commands send to the set up the scope: HEAD OFF; DIS:GRA GRID; DIS:INTENSIT:WAVE 80; HOR:RECORDL 10000; ACQ:MODE SAM; ACQ:STOPA SEQ; MEASU:GAT SCRE; SEL:CH1 ON; C...
by jaime_g
March 3rd, 2017, 10:55 am
Forum: 2600 Series SourceMeter
Topic: 2636B generate current single pulse
Replies: 4
Views: 15458

Re: 2636B generate current single pulse

I reviewed the code and found some errors with the measurement timer configuration. :( Here is an improved version: reset() display.screen = display.SMUA display.smua.digits = display.DIGITS_4_5 display.smua.measure.func = display.MEASURE_DCVOLTS smua.sense = smua.SENSE_REMOTE smua.source.func = smu...
by jaime_g
March 1st, 2017, 2:59 pm
Forum: 2600 Series SourceMeter
Topic: Trigger Configuration to Measure CE Sat Voltage
Replies: 0
Views: 6445

Trigger Configuration to Measure CE Sat Voltage

I am trying to measure the Collector-Emitter Saturation Voltage of an NPN transistor. The Vce drops from 10V to ~0.2V when the transistor is in its saturated state. To achieve saturation a 1.55mA, 500µs pulse is applied to the base terminal with the collector bias voltage set at 10V. An SMU21612B is...
by jaime_g
February 28th, 2017, 3:23 pm
Forum: 2600 Series SourceMeter
Topic: 2636B generate current single pulse
Replies: 4
Views: 15458

Re: 2636B generate current single pulse

I noticed the number of points in the PulseSweepISingle is set to 11. For a single pulse this should be set to 1. Here is a script I use for a single 1ms, 10mA pulse using channel B: reset() display.screen = display.SMUB display.smub.digits = display.DIGITS_4_5 smub.sense = smub.SENSE_REMOTE smub.me...
by jaime_g
February 27th, 2017, 11:55 am
Forum: 2600 Series SourceMeter
Topic: 2612B Single Pulse Programming
Replies: 4
Views: 16607

Re: 2612B Single Pulse Programming

Thanks Andrea. Your recommendations made all the difference.

Jaime Arevalo
by jaime_g
February 27th, 2017, 10:41 am
Forum: 2600 Series SourceMeter
Topic: 500µs 10mA Pulse for CE breakdown voltage
Replies: 3
Views: 13348

Re: 500µs 10mA Pulse for CE breakdown voltage

Hi Andrea, Thanks for your prompt response. The short answer to your query is no. The SMU2612B channel B force and sense leads were connected to the interface test adapter (ITA). However, disregard the code from the post above, it was an oversight on my side. I reviewed the programming commands and ...
by jaime_g
February 24th, 2017, 3:40 pm
Forum: 2600 Series SourceMeter
Topic: 500µs 10mA Pulse for CE breakdown voltage
Replies: 3
Views: 13348

Re: 500µs 10mA Pulse for CE breakdown voltage

SMU2612B 50V-500µs Pulse.png
SMU2612B 50V-500µs Pulse.png (7.11 KiB) Viewed 13339 times
by jaime_g
February 24th, 2017, 11:15 am
Forum: 2600 Series SourceMeter
Topic: 500µs 10mA Pulse for CE breakdown voltage
Replies: 3
Views: 13348

500µs 10mA Pulse for CE breakdown voltage

Dear Forum, Has anyone tried to configure an SMU2612B to generate a 500µs, 10mA pulse? I have been trying to configure an SMU2612B (with v3.2.2 firmware) to generate a 500µs, 10mA, (55V limit) pulse to measure the CE breakdown voltage of NPN transistor. Maybe I am missing something trivial on how to...