Search found 109 matches
- April 24th, 2013, 1:15 pm
- Forum: Other Power Supplies
- Topic: Model 248 Earth Ground
- Replies: 1
- Views: 11827
Re: Model 248 Earth Ground
Hello, The 248 has an OVP circuitry to avoid arcing. This is explained in the datasheet (Model 248 High Voltage Supply and Technical Information - DC Power Supplies) See link below; http://www.keithley.com/support/keidoc_searchresult?keyword=248&item_type=Data+Sheet In addition, the E-handbook there...
- April 24th, 2013, 12:56 pm
- Forum: 2651A High Current SourceMeter
- Topic: Using 2651A at Higher DC current
- Replies: 4
- Views: 20430
Re: Using 2651A at Higher DC current
Hi Jerry,
You can parallel 2x2651A to meet your requirements. In TSP Express (built in software), there's an option for you to combine the SMU.
It is a Download Java Runtime J2SE(TM) Environment.
http://www.keithley.com/support/keidoc_ ... e=Software
Jane
You can parallel 2x2651A to meet your requirements. In TSP Express (built in software), there's an option for you to combine the SMU.
It is a Download Java Runtime J2SE(TM) Environment.
http://www.keithley.com/support/keidoc_ ... e=Software
Jane
- April 19th, 2013, 6:41 pm
- Forum: 2600 Series SourceMeter
- Topic: 2602A, Clearing memory which is not set up as a buffer
- Replies: 3
- Views: 16014
Re: 2602A, Clearing memory which is not set up as a buffer
Hello, First of all, the current firmware for the 260xA SMU is in the link below. http://www.keithley.com/support/keidoc_searchresult?keyword=2602A&item_type=Firmware You might want to upgrade your unit. Secondly, since you didn't mention what kind of test you are running, I am going to direct you t...
- April 19th, 2013, 6:06 pm
- Forum: Series 2000 and Series 2100 DMM's
- Topic: 2110 DMM- Therocouple reading max speed
- Replies: 3
- Views: 16274
Re: 2110 DMM- Therocouple reading max speed
Hello, Here's the speed specification of 2110. Speed At 5½ digits, the Model 2110 delivers up to 200 readings/s via the USB remote interface. At the fast 4½-digit setting, it reads up to 50,000 readings/s and up to 30,000 readings/s into the buffer, making it ideal for production and monitoring appl...
- April 19th, 2013, 5:41 pm
- Forum: S530 Parametric Test Systems
- Topic: Triax or Quadrax connectors on S400 Relay Matrix card
- Replies: 1
- Views: 12552
Re: Triax or Quadrax connectors on S400 Relay Matrix card
Hello,
Please check page 5 of the attached file. The attached file is our cable selector guide.
I hope this helps.
Jane
Please check page 5 of the attached file. The attached file is our cable selector guide.
I hope this helps.
Jane
- April 17th, 2013, 4:45 pm
- Forum: 2600 Series SourceMeter
- Topic: Low current measurement 2634B
- Replies: 1
- Views: 10050
Re: Low current measurement 2634B
Hello,
The attached file is the low current measurement application note. In the last page, it talks about the specifics of dark current measurement including set-up and test fixtures.
Jane
The attached file is the low current measurement application note. In the last page, it talks about the specifics of dark current measurement including set-up and test fixtures.
Jane
- April 17th, 2013, 11:51 am
- Forum: 2400 Series SourceMeter
- Topic: Measure the voltage of CNTs
- Replies: 7
- Views: 27092
Re: Measure the voltage of CNTs
Hi bear2012, The available freeware that we have for 2410 is the link below. http://www.keithley.com/support/keidoc_searchresult?keyword=2410&item_type=Software Keep in mind that this software is not a supported software. Otherwise, the alternative is the link below. We only post its documentation. ...
- April 16th, 2013, 9:19 pm
- Forum: 2400 Series SourceMeter
- Topic: Keithley 2400 surge current
- Replies: 2
- Views: 13234
Re: Keithley 2400 surge current
Hello,
There's a possibility that during your power up cycle/sequence of the instrument a transient happens, which could be higher than 0.1V. Is there a way to debug your test(s) using a scope maybe to see whether there's a spike going on?
Jane
There's a possibility that during your power up cycle/sequence of the instrument a transient happens, which could be higher than 0.1V. Is there a way to debug your test(s) using a scope maybe to see whether there's a spike going on?
Jane
- April 16th, 2013, 9:05 pm
- Forum: 4200-SCS Semiconductor Parameter Analyzer
- Topic: KTEIv8.2 Installation on 4200-SCS System
- Replies: 2
- Views: 13554
Re: KTEIv8.2 Installation on 4200-SCS System
Hi Radu, It is not clear what else changed after you re-installed Windows to the system. Who knows? Something could have been changed before installation. Why do you have to re-install Windows? I think it is best to contact the nearest Keithley 4200 service office to you. At times, starting from scr...
- April 16th, 2013, 8:48 pm
- Forum: 2600 Series SourceMeter
- Topic: 2612 incorrectly measuring current under 1mA range
- Replies: 6
- Views: 24709
Re: 2612 incorrectly measuring current under 1mA range
Check the response after the unit is back from repair if the range change during the 1000 pts sweep has some effect assuming you use autorange with reasonable delay between points.
- April 16th, 2013, 8:33 pm
- Forum: 2600 Series SourceMeter
- Topic: Remote Control Software
- Replies: 3
- Views: 16481
Re: Remote Control Software
Hello,
You can use our TSB for your project.
See the link below on where to download.
http://www.keithley.com/support/keidoc_ ... e=Software
Jane
You can use our TSB for your project.
See the link below on where to download.
http://www.keithley.com/support/keidoc_ ... e=Software
Jane
- April 16th, 2013, 8:28 pm
- Forum: 2600 Series SourceMeter
- Topic: 2602A USB problem
- Replies: 6
- Views: 24856
- April 16th, 2013, 7:16 pm
- Forum: Electrometers
- Topic: Help on 8009 test fixture for electrometer 6517B
- Replies: 1
- Views: 10608
Re: Help on 8009 test fixture for electrometer 6517B
Hello, You are correct. The design of 8009 test fixture is different than an AC or two port type of measurement. The interconnect/cabling are totally different between CV and IV application. I am not sure why you wanted to use the 8009. Why not use a probe? With our 4200-SCS parameter analyzer, you ...
- April 16th, 2013, 7:01 pm
- Forum: Other DMM's
- Topic: Keithley 2500 Dual Photodiode Meter firmware
- Replies: 1
- Views: 15219
Re: Keithley 2500 Dual Photodiode Meter firmware
Hello, Unfortunately, the Model 2500 & 2502 have no new firmware. I assume that the other 9 units have the same FW as the one that is problematic. If so, it might be best to contact our Service organization and request for RMA. Send an email to [email protected].com They will send you instruction...
- April 16th, 2013, 6:51 pm
- Forum: 2600 Series SourceMeter
- Topic: 2602a unable to connect to LAN
- Replies: 1
- Views: 10513
Re: 2602a unable to connect to LAN
Hello,
See attached instruction for LAN Configuration. See page 281 - 310 for your reference.
Let us know how it goes.
See attached instruction for LAN Configuration. See page 281 - 310 for your reference.
Let us know how it goes.
- April 16th, 2013, 6:32 pm
- Forum: 2600 Series SourceMeter
- Topic: Elearning TSB broken link
- Replies: 4
- Views: 19000
Re: Elearning TSB broken link
Hello,
Thank you for bringing this to our attention. I informed our training personnel regarding the matter.
Thank you,
Jane
Thank you for bringing this to our attention. I informed our training personnel regarding the matter.
Thank you,
Jane
- February 27th, 2013, 7:09 pm
- Forum: 2600 Series SourceMeter
- Topic: Unwanted voltage between the SENSE contacts
- Replies: 4
- Views: 18767
Re: Unwanted voltage between the SENSE contacts
Hello, I believed you also sent a similar concern regarding your compliance measurement. I believed it is a simalar issue with what I replied to you. I think that the issue is more of a set-up issue than the measurement itself. Since you have a 2636A, you can actually use the TSP Express to do your ...
- February 27th, 2013, 7:07 pm
- Forum: 2600 Series SourceMeter
- Topic: 2636A measurement during compliance
- Replies: 1
- Views: 10089
Re: 2636A measurement during compliance
Hello, I think that the issue is more of a set-up issue than the measurement itself. Since you have a 2636A, you can actually use the TSP Express to do your specific measurement and settings. Another aspect of the TSP Express is that it can generate the code for you as well. Use this as your debuggi...
- February 15th, 2013, 7:04 pm
- Forum: 2600 Series SourceMeter
- Topic: Keithley 2612A activex, dot.net or alternative
- Replies: 1
- Views: 10111
Re: Keithley 2612A activex, dot.net or alternative
Hello,
What you are asking for has never been done before. What might work better for you is to read the data from the saved data *.csv file into your LV program. At this point, you now have multiple options on how to port your data to LV.
What you are asking for has never been done before. What might work better for you is to read the data from the saved data *.csv file into your LV program. At this point, you now have multiple options on how to port your data to LV.
- February 15th, 2013, 3:43 pm
- Forum: 2600 Series SourceMeter
- Topic: Display wrong current
- Replies: 3
- Views: 15838
Re: Display wrong current
Hello, Please list your 2600 SMU test conditions, source and measure stimulus per channel, source and measure range on SMU, how you connect to your device, 2W/4W. What is the capacitive load of your device and any other pertinent information. Also the attached document below, differentiate a DMM fro...