Re: TPP0201 PROBE compensation capacitor?

It is an older design scope, but is still a model that is sold new:
https://www.tek.com/oscilloscope/tds200 ... cilloscope
- so it is definitely more than adequate for a beginner! Congratulations!
by w2aew
July 29th, 2019, 5:29 am
Forum: Other or Discontinued Oscilloscopes
Topic: TPP0201 PROBE compensation capacitor?
Replies: 3
Views: 1240

Re: TPP0201 PROBE compensation capacitor?

The compensation capacitor on the TPP0201 probe is located just behind the BNC connector, immediately underneath where the "TPP0201" model number is printed.
by w2aew
July 3rd, 2019, 5:42 am
Forum: Older and Obsolete Spectrum Analyzers
Topic: RSA3303A operating system / firmware reinstall
Replies: 3
Views: 1616

Re: RSA3303A operating system / firmware reinstall

It is possible that you have a corrupted copy of the firmware installation files. The latest/last version of firmware can be found here: https://www.tek.com/wca230a-software/rsa3000b-and-rsa3000a-series-firmware-v420009 The zip file should show a file size of 42,710KB in size. I have a zip file that...
by w2aew
June 18th, 2019, 6:06 am
Forum: RSA Programming
Topic: WLAN Measurements
Replies: 1
Views: 1625

Re: WLAN Measurements

You will have to have a license for one (or more) of the WLAN Analysis options, depending upon the particular version of the standard you're testing. Option SV23 covers 802.11a/b/g/j/p, option SV24 adds 802.11n, and option SV25 adds 801.11ac. The licenses are cumulative, which means that if you need...
by w2aew
May 30th, 2019, 9:18 am
Forum: Non- Windows Based Oscilloscopes
Topic: MSO4104 time/div is wrong
Replies: 3
Views: 2820

Re: MSO4104 time/div is wrong

OK, it wasn't clear that you validated the frequency of the compensation signal with a second device.

This is something that should be able to be adjusted/repaired with a fresh scope calibration.
by w2aew
May 30th, 2019, 8:38 am
Forum: Non- Windows Based Oscilloscopes
Topic: MSO4104 time/div is wrong
Replies: 3
Views: 2820

Re: MSO4104 time/div is wrong

The 1kHz calibration signal on the scope is *not* intended to be a calibrated 1kHz signal source. It's only function is to be used to *compensate* your 10x probes. It is more likely that the 1kHz compensation signal has changed, and not the scope's horizontal calibration. You should be able to valid...
by w2aew
May 24th, 2019, 11:38 am
Forum: RSA300/RSA500/RSA600 Series
Topic: Ref In Coupling
Replies: 5
Views: 4166

Re: Ref In Coupling

Engineering has advised that the Maximum DC voltage that can be applied to the Reference Input is 10V.
by w2aew
May 24th, 2019, 9:10 am
Forum: RSA300/RSA500/RSA600 Series
Topic: Ref In Coupling
Replies: 5
Views: 4166

Re: Ref In Coupling

I'm working on getting that number. Likely safe to say that it will be 25Vdc or greater, maybe as much as 40-50V. As soon as i get confirmation, I'll post it.
by w2aew
May 21st, 2019, 12:58 pm
Forum: RSA300/RSA500/RSA600 Series
Topic: Ref In Coupling
Replies: 5
Views: 4166

Re: Ref In Coupling

It is AC-coupled, 50 ohms.
by w2aew
April 27th, 2019, 5:05 pm
Forum: Other or Discontinued Oscilloscopes
Topic: Need IC for scope
Replies: 1
Views: 2403

Re: Need IC for scope

This is a Tektronix custom IC, designed and fabricated in Tek’s own integrated circuit fab in the early-mid 1970s. Here is a link to its datasheet: http://w140.com/tekwiki/images/6/61/Tek_155-0050-01.pdf The only way to get a replacement is from a donor scope, or from a vendor that sells new-old-sto...
by w2aew
April 4th, 2019, 1:26 pm
Forum: General Discussion
Topic: sample interval
Replies: 1
Views: 3056

Re: sample interval

The sample interval will be a function of two things: The horizontal scale setting (time/div), and the Record Length that you have configured. By default, the Record length is 10,000 samples. If you need a smaller sample interval at a given time/div setting, you would go into the Acquire menu and in...
by w2aew
January 17th, 2019, 1:53 pm
Forum: AWG5000/7000/70000 Series
Topic: AWG70002A Maximum Output Frequency
Replies: 2
Views: 4549

Re: AWG70002A Maximum Output Frequency

Using the Basic Waveform Plug-In, you can create a 6.4GHz sinewave. This gives you an oversample rate of 2.5x, which is the minimum recommended for RF applications. The other RF plug-ins can also be used to create more complex RF waveforms with carrier frequencies up to 6.4GHz.
by w2aew
December 30th, 2018, 5:49 pm
Forum: Older and Obsolete Spectrum Analyzers
Topic: Tek 3066
Replies: 3
Views: 5332

Re: Tek 3066

Ok. I can’t make any promises. The 3066/86 pre-dates my time with Tektronix by more than a decade.
by w2aew
December 27th, 2018, 7:10 pm
Forum: Older and Obsolete Spectrum Analyzers
Topic: Tek 3066
Replies: 3
Views: 5332

Re: Tek 3066

I'm not sure if these will help you - but take a look. Here are the images from the disks for the 3086 (same as the 3066 but includes wideband IF and IQ inputs).

https://fortive.box.com/s/lqzhtaimh74gp ... 4gfgrveto8
by w2aew
December 12th, 2018, 7:19 am
Forum: RSA Programming
Topic: Setting the Acquisition Sampling Freqency
Replies: 4
Views: 6774

Re: Setting the Acquisition Sampling Freqency

The hardware in the instrument controls the IQ sample rate as a function of the acquisition BW. There is a table indicating the available acquisition BW settings and the associated IQ sample rates (as well as the recording depth and timing resolution) at each setting. It is on page 33 of the datashe...
by w2aew
December 11th, 2018, 3:59 pm
Forum: RSA Programming
Topic: Setting the Acquisition Sampling Freqency
Replies: 4
Views: 6774

Re: Setting the Acquisition Sampling Freqency

The Acquisition Sampling Frequency is intimately tied to the Acquisition Bandwidth. It is not a function of RBW or acquisition length. There are specific, discrete values of Acq BW and Sample Rate. In other words, you can't set a specific user-defined value. It will always round "up" the the nearest...
by w2aew
December 3rd, 2018, 7:19 am
Forum: General Discussion
Topic: Press 'Single' button on oscilloscope programmatically
Replies: 2
Views: 5704

Re: Press 'Single' button on oscilloscope programmatically

First issue an ACQUIRE: STOPAFTER SEQUENCE
This will setup the Single acquisition.

Then, ACQUIRE:STATE ON will put the scope in Run mode, waiting for the trigger event.

Once the scope triggers, it will be stopped.

To re-arm the Single mode, issue ACQUIRE:STATE ON again...
by w2aew
November 20th, 2018, 10:48 am
Forum: Non- Windows Based Oscilloscopes
Topic: MDO 3102 input capacitance
Replies: 1
Views: 3901

Re: MDO 3102 input capacitance

The front panel on mine indicates 13pF for the 1Mohm input termination.
by w2aew
August 30th, 2018, 1:38 pm
Forum: RSA300/RSA500/RSA600 Series
Topic: rsa306 - Error 302 when setting rbw to 8
Replies: 4
Views: 8683

Re: rsa306 - Error 302 when setting rbw to 8

SignalVu-PC allows you to set RBW down to 1Hz when the Span is <=1MHz.
by w2aew
August 14th, 2018, 1:13 pm
Forum: RSA300/RSA500/RSA600 Series
Topic: rsa306 - Error 302 when setting rbw to 8
Replies: 4
Views: 8683

Re: rsa306 - Error 302 when setting rbw to 8

From the API manual (page 88), users can use SPECTRUM_GetLimits to get the limit of the spectrum settings. The limit table shows the minimum rbw can be set using API is 10 Hz.