Re: Keithley Engineering Notes

Hello, What in specific do you need? I am not sure we have this older notes.
by Dale C
May 30th, 2019, 10:48 am
Forum: PicoAmmeters
Topic: keithley 6487 measuring current at 5 kV
Replies: 1
Views: 2124

Re: keithley 6487 measuring current at 5 kV

do you need an external source of 5kV? Or do you need to float the 6487 at 5kV? Model 6487 has its own voltage source built in. With a max output of 500V. But if you are testing at DUT at 5kV, then make sure you do not float the 6487 at more than 500V. Attached is a document showing a protection cir...
by Dale C
April 25th, 2019, 8:04 am
Forum: Series 2200 Programmable DC Power Supplies
Topic: Response to load changes on Model 2200 power supplies.
Replies: 0
Views: 2271

Response to load changes on Model 2200 power supplies.

The issue is the 2200 supply channel control circuit reacts quickly to the very large load current change. For example... 10mA to an 1A. That is a factor of 100. All supplies will have a transient response to the large load change. Then the supply will take some time to recover. Reducing the source ...
by Dale C
April 24th, 2019, 8:16 am
Forum: 2400 Series SourceMeter
Topic: Windows10 driver for KUSB-488B USB to GPIB
Replies: 71
Views: 184313

Re: Windows10 driver for KUSB-488B USB to GPIB

Hello Diogo,
Please call 503-627-2445/
that is the service department in Oregon.
They can help you return the unit for service.
Or go to this link.
https://www.tek.com/repair


Hope this helps.
Dale Cigoy
by Dale C
March 26th, 2019, 7:52 am
Forum: Series 2000 and Series 2100 DMM's
Topic: KEITHLY 2100 100MOhm range adjustment
Replies: 1
Views: 2769

Re: KEITHLY 2100 100MOhm range adjustment

100M range calibration adjustment. Steps 31 and 32 need to be performed. This is usually the issue. Need to perform steps 31 and 32 in the calibration procedure. Basically, you do not use the actual calibrator output. You use 99.99750 no matter what the calibrator output is. And the calibrator is in...
by Dale C
March 12th, 2019, 10:35 am
Forum: Electrometers
Topic: 6514 over RS-232 and ExceLINX driver problem
Replies: 8
Views: 23839

Re: 6514 over RS-232 and ExceLINX driver problem

I would go with the WIN 7 machine, if you can.
And the KUSB-488A is not supported in WIN 10.
Also the Kickstart software does not yet support the Model 6514.
Software could be the Labview driver as it has some examples. But you need Labview.
by Dale C
March 11th, 2019, 5:18 am
Forum: Other Keithley Instrument Products
Topic: Keithley 2182a programming
Replies: 2
Views: 4347

Re: Keithley 2182a programming

Hello Reed, I used this SCPI command sequence and then applied 37 separate trigger pulses. Then got the 37 measurements from the buffer. *RST :INIT:CONT OFF :SENSE:FUNC 'VOLT' :SENSE:VOLT:RANGE 0.01 :VOLT:NPLC 0.1 :TRIG:SOURCE EXT :TRIG:COUNT 37 :TRIG:DEL 0.0 :TRAC:POINTS 37 :TRAC:FEED SENSE1 :TRAC:...
by Dale C
February 22nd, 2019, 10:01 am
Forum: Other DMM's
Topic: 2019 DMM4050 Newest Firmware verification
Replies: 12
Views: 11606

Re: 2019 DMM4050 Newest Firmware verification

Looks like 2.31 is the latest firmware revision for the DMM4040/4050.
This is of 2-22-19.
by Dale C
February 19th, 2019, 10:11 am
Forum: Series 2700 Multimeter/Data Aquisition/Switch System
Topic: HELP 2700 + 7700
Replies: 1
Views: 3435

Re: HELP 2700 + 7700

Looks like you have to have a command to encompass the entire scan list.
Like this.
:ROUTE:SCAN (@101:115)
That should do it.
by Dale C
February 13th, 2019, 6:50 am
Forum: KickStart
Topic: KICKSTARTFL-HRMA Manual
Replies: 1
Views: 3710

Re: KICKSTARTFL-HRMA Manual

Sorry, there is no separate manual for KICKSTART-HRMA.
The help system in Kickstart is all that is available.
What specific questions do you have regarding KICKSTART -HRMA?
by Dale C
February 1st, 2019, 7:31 am
Forum: Other Source Measurement Units (SMU)
Topic: Append voltage sweep in kei 236
Replies: 2
Views: 4810

Re: Append voltage sweep in kei 236

I ran a crude Labview 2016 VI to do this. It is attached. But here are the DDC command codes to show that Q7 works. F0,1X Source Volt measure Current B0,0X Set Bias level to 0V Q1,1,10,0,500X Set linear stair sweep from 1V to 10V in 1V steps with 0.5seconds delays Q7,2,4,1,0,500X Append linear start...
by Dale C
February 1st, 2019, 5:46 am
Forum: Low Current Measurements
Topic: Effective input resistance in current mode at high temp
Replies: 2
Views: 4914

Re: Effective input resistance in current mode at high temp

The effect on input resistance is not so much the ambient temperature as the internal temperature on the unit(6514). As long as the unit(6514) is warmed up, usually about one hour, the input resistance should not be affected much. Here is the input resistance calculation of the Model 6487 in current...
by Dale C
January 25th, 2019, 7:31 am
Forum: Electrometers
Topic: Keithley 6517a to computer connection
Replies: 5
Views: 19166

Re: Keithley 6517a to computer connection

Just use the VI that is in the above communication... 6517BRead.VI
That should do it. I could do this with Labview 2016.
by Dale C
January 24th, 2019, 8:47 am
Forum: Series 2700 Multimeter/Data Aquisition/Switch System
Topic: Advanced Scan Syntax
Replies: 1
Views: 3410

Re: Advanced Scan Syntax

One of the errors is four wire measurement on slot two at channels 11 and 12. This is the 7700 card with only 20 two pole channels. So channel 11 would pair channel 21 when there is no channel 21. That is probably the -221 error. The -104 error is a syntax error. I noticed that the command :FUNC TEM...
by Dale C
January 22nd, 2019, 11:42 am
Forum: General Discussion
Topic: taking 100 DCV measurements with constant duration using Keithley 2700
Replies: 2
Views: 5117

Re: taking 100 DCV measurements with constant duration using Keithley 2700

Start off the program sequence of commands with a
*RST
Then when ready to take data send the
:INIT
command.
Then when you want the data send
:TRACE:DATA?
That should send back all 100 points of measurement stored in the buffer when addressed to talk.
Hope this helps.
by Dale C
January 22nd, 2019, 11:28 am
Forum: Series 2200 Programmable DC Power Supplies
Topic: 2230G-30-1 List Programming
Replies: 1
Views: 3745

Re: 2230G-30-1 List Programming

Sorry, only the single channel units have the LIST mode.
The SHIFT that is referred to is only on the single channel units.
You could always use the USB or GPIB port and create a list in a loop in software...
by Dale C
January 22nd, 2019, 10:49 am
Forum: 2400 Series SourceMeter
Topic: Keithley 2410 sweep time much longer than it should be
Replies: 1
Views: 4577

Re: Keithley 2410 sweep time much longer than it should be

There are a number of parameters that affect the system speed. 1. NPLC- Number of Power Line Cycles. This is only part of the A/D conversion process. So 1PLC does not mean it will be 16.67msec per reading. 2. Ranging - Auto range can cause delays. For more precise timing use fixed range. 3. Auto Zer...
by Dale C
January 22nd, 2019, 10:42 am
Forum: 2600 Series SourceMeter
Topic: TSP Express Unable to Launch
Replies: 4
Views: 5388

Re: TSP Express Unable to Launch

Try this... Cannot open or access TSP Express with 2600B series SMU. What internet browser are you using? Looks like Firefox is the only one that supports Java. I think that might be the issue. Firefox 52 ESR is available. Unlike the regular Firefox, the ESR (Extended Support Release) series will no...
by Dale C
January 22nd, 2019, 10:37 am
Forum: Series 2000 and Series 2100 DMM's
Topic: Keithley 2000 measuring real temperature slow
Replies: 1
Views: 3515

Re: Keithley 2000 measuring real temperature slow

sounds like you are using the SIMulated reference junction on the Model 2000 for thermocouples. Suggestion... If you do not have a Model 2001TCSAN card with a CJC for the reference junction, I would suggest to let the Model 2000 warm up for one hour and use the SIM reference temperature of 27 or 28C...
by Dale C
January 3rd, 2019, 8:14 am
Forum: Electrometers
Topic: compatibility issue
Replies: 1
Views: 4049

Re: compatibility issue

Is Model 6517A set for GPIB? If so, what address?(default is 27).
Since you have NI-VISA, you should have NI-MAX.
Use NI-MIX to see if you can control the 6517A with *IDN?