by brian.d.smith
February 13th, 2017, 6:14 am
Forum: Series 2700 Multimeter/Data Aquisition/Switch System
Topic: 2700 not seeing 7703 module
Replies: 8
Views: 29602

Re: 2700 not seeing 7703 module

I determined there is a checksum error in the upper portion of the 7703 eeprom. The checksummed range is from 0x10 to 0x66. If I correct the checksum the card is recognized but doesn't work properly. I can't toggle any relays. Does anyone have an example of a 7703 card eeprom? I'd really like to us...
by brian.d.smith
February 8th, 2017, 5:33 am
Forum: Series 2700 Multimeter/Data Aquisition/Switch System
Topic: 2700 not seeing 7703 module
Replies: 8
Views: 29602

Re: 2700 not seeing 7703 module

For the 7703, can you open and close the relays from the front panel of the 2700?
by brian.d.smith
February 6th, 2017, 3:41 pm
Forum: 2600 Series SourceMeter
Topic: Current limit
Replies: 1
Views: 9857

Re: Current limit

You can set a limit for current. Press the LIMIT button the front panel and navigate to the menu to set current limit.

The instrument will not "shut off" automatically. You can create a script that will turn off the output if the current is greater than or equal to a value that you specify.
by brian.d.smith
February 2nd, 2017, 9:19 am
Forum: 2400 Series SourceMeter
Topic: 2400-voltage measurement before and after current pulse
Replies: 2
Views: 13036

Re: 2400-voltage measurement before and after current pulse

If you want to measure the characteristics of the rising and falling edge of the pulse from a 2400, you will need to use an oscilloscope.

Your other option is to use a sourcemeter with a high-speed digitizer. The 2461 has a high speed digitizer.
by brian.d.smith
February 2nd, 2017, 9:14 am
Forum: 2450/2460 Touchscreen SourceMeter
Topic: Bugs detected in 2450
Replies: 5
Views: 22576

Re: Bugs detected in 2450

zvikae wrote:Hi brian.
I have the same problem case #3, 2450 freezes after changing GRAPH settings.
please advise.
What version of firmware is your 2450 running?
by brian.d.smith
January 25th, 2017, 7:20 am
Forum: 2400 Series SourceMeter
Topic: 2400 firmware
Replies: 2
Views: 10503

Re: 2400 firmware

Your email address is not listed in your forum profile. Please add your email address to your profile and I can send the firmware to you.
by brian.d.smith
January 9th, 2017, 10:15 am
Forum: Series 2000 and Series 2100 DMM's
Topic: Continuous voltage measurement with Matlab script, 2000
Replies: 8
Views: 30693

Re: Continuous voltage measurement with Matlab script, 2000

Hello Andrea, Thanks for your help. Just one more problem, I added these lines to the script fprintf(dev,'*RST'); fprintf(dev, ':CONF:VOLT:DC'); pause(0.1); fprintf(dev, 'SENS:VOLT:DC:NPLC:0.1'); anyhow despite the fact that I change the rate to fast and manual measurement in the multi-meter, whene...
by brian.d.smith
January 9th, 2017, 6:13 am
Forum: Series 3700 Switch System/Multimeter and Plug-In Cards
Topic: Keithley 3706A cannot save Setup
Replies: 8
Views: 30473

Re: Keithley 3706A cannot save Setup

Attached is the Python program. Change the .txt extension to a .py extension.
by brian.d.smith
January 9th, 2017, 5:41 am
Forum: Series 3700 Switch System/Multimeter and Plug-In Cards
Topic: Keithley 3706A cannot save Setup
Replies: 8
Views: 30473

Re: Keithley 3706A cannot save Setup

Attached is a Python program that scans two wire ohms with one MUX card # This python program controls a 3706A DMM with a 3720 switch card and a 3720-ST brakout board. Nine resistors are installed in channels # 1 - 9. Two-wire ohms mesurements are scanned. The strings that are returned by the 3706A ...
by brian.d.smith
January 8th, 2017, 1:01 pm
Forum: Series 2000 and Series 2100 DMM's
Topic: Continuous voltage measurement with Matlab script, 2000
Replies: 8
Views: 30693

Re: Continuous voltage measurement with Matlab script, 2000

Your script appears to be lacking some commands. The script setup is only fprintf(dev, ':CONF:VOLT:DC'); pause(0.1); I recommend that you begin with a *rst After CONF:VOLT:DC, I would set SENS:VOLT:DC:NPLC to the NPLC setting you are using for front panel measurements. I believe that the default spe...
by brian.d.smith
January 5th, 2017, 5:11 am
Forum: Series 2000 and Series 2100 DMM's
Topic: Continuous voltage measurement with Matlab script, 2000
Replies: 8
Views: 30693

Re: Continuous voltage measurement with Matlab script, 2000

Can you post the actual script that Matlab is running?
by brian.d.smith
December 26th, 2016, 10:23 pm
Forum: Low Voltage Measurements
Topic: Driver for Nanovoltmeter 2182A
Replies: 1
Views: 12708

Re: Driver for Nanovoltmeter 2182A

There appears to be no driver for the A version. The recommendation is to use the driver for the non-A version. There are two releases on tek.com http://www.tek.com/specialty-instruments/2182-software/2182-labview-ver-5x-lv-driver-windows http://www.tek.com/specialty-instruments/2182-software/2182-l...
by brian.d.smith
December 2nd, 2016, 1:24 am
Forum: 2400 Series SourceMeter
Topic: Keithley 2410 sourcemeter blue LED always on
Replies: 1
Views: 10554

Re: Keithley 2410 sourcemeter blue LED always on

You should contact our service organization to have your 2410 repaired. The blue light should be extinguished when you depress the output on/off button.
by brian.d.smith
December 2nd, 2016, 1:19 am
Forum: 2600 Series SourceMeter
Topic: 2612B Synchronization Method
Replies: 3
Views: 16144

Re: 2612B Synchronization Method

It is not clear to me what you are trying to synchronize and why. Are you performing parallel test of multiple devices on a wafer? Do you have a mix of 2600A and 2600B instruments?

Please explain your test setup and procedures in detail.
by brian.d.smith
December 1st, 2016, 2:05 am
Forum: 2400 Series SourceMeter
Topic: Keithley 2400 Sub-nA Measurements with Triax Cable
Replies: 4
Views: 19489

Re: Keithley 2400 Sub-nA Measurements with Triax Cable

The guard protects against current leakage in the cable between the instrument and your circuit. The guard is dropped just before your circuit. The guard can be used to energize an enclosure that surrounds your circuit. This is called a guard enclosure. It prevents current from flowing from HI to LO...
by brian.d.smith
November 30th, 2016, 6:22 pm
Forum: 2400 Series SourceMeter
Topic: Keithley 2400 Sub-nA Measurements with Triax Cable
Replies: 4
Views: 19489

Re: Keithley 2400 Sub-nA Measurements with Triax Cable

You would have to use the rear banana jacks and connect the banana jack guard to the guard of a triax cable. The model 2400 range uncertainty of the lowest current range is 300pA. This may preclude any serious sub nano-amp work. The 2450 is a much better platform for sub nano-amp work. I have perfor...
by brian.d.smith
November 29th, 2016, 10:58 pm
Forum: 2400 Series SourceMeter
Topic: Measure Only feature of 2410
Replies: 4
Views: 19072

Re: Measure Only feature of 2410

The 2410 is not an AC volts instrument. If you want to measure AC volts, you want to use a DMM.
by brian.d.smith
November 4th, 2016, 2:11 pm
Forum: 2657A High Voltage SourceMeter
Topic: SMU 2657A overheating?
Replies: 5
Views: 23089

Re: SMU 2657A overheating?

If you are never exceeding 50uA at 3000V and below, you should never have a heat problem. My guess is that the 2657A is oscillating and that the oscillation is what is heating up the instrument. Can you get a high voltage probe and record the sweeps with an oscilloscope? Is your device highly capaci...
by brian.d.smith
November 2nd, 2016, 4:53 am
Forum: 2400 Series SourceMeter
Topic: Keithley 2400
Replies: 2
Views: 13821

Re: Keithley 2400

From the 2011 Users Manual page 4-14 verbatim. Please have a look and read the description of each instruction. *RST FUNC “RES” RES:MODE AUTO RES:RANG 20E3 :SYST:RSEN ON :FORM:ELEM RES :OUTP ON :READ? :OUTP OFF The above script uses 4-wire so you will want to change the sense mode *RST FUNC “RES” RE...