Hi,
I am doing mixed IV and VI test and I need fast measurment speed and high accuracy. I found list measurment can do only single type of scan: IV or VI. Can Keithley 2400 Series do mixed type scan? If not, can any other Keithley series do mixed type scan? Thank you very much.
Best,
Aki
Nathan S
Keithley Applications
Contact:
Re: Can 2400 do mixed scan?
Post by Nathan S » December 19th, 2011, 9:13 am
Aki,
Yes the 2400 can do mixed Source I and Source V sweeps. I would take a look at Section 10 in the 2400 User's Manual which is all about sweeping. In there, look for the fourth type of sweep called "Source Memory sweep." You get up to 100 points where you can change most parameters on a point by point basis, including the source function. Below is a link to download that manual.
http://www.keithley.com/support/data?asset=887
akinokoe
Post by akinokoe » December 19th, 2011, 9:21 am
Oh Great! Thank you very much for your help Nathan