:READ? return in simultaneous voltage and current measurement
JeanSu
Country: Switzerland
Post by JeanSu » January 11th, 2019, 4:39 am
Hi everyone,
I have a Keithley 2410, which I address using my own Python program. I am applying a fixed voltage and want to measure the real voltage (not the set one) and the current. The problem is that I do not understand the returned values from the :READ? command.
WRITE: ":SOUR:FUNC VOLT"
WRITE: ":SOUR:VOLT:RANG:AUTO 0;:SOUR:VOLT:RANG 20"
WRITE: ":SOUR:VOLT:LEV 0.5"
WRITE: ":SENS:CURR:PROT 1
WRITE: ":FUNC 'VOLT', 'CURR'"
WRITE: "OUTPUT ON"
WRITE: ":READ?"
Return value: +4.998975E-01 (the applied 0.5V),+4.992652E-03 (current over 100Ohm resistor) ,+9.910000E+37 (?),+1.346539E+04(?),+2.355600E+04(?)
Does anybody know what these values stand for or where I can read about the return value? I already looked through the UserManual but didn't find anything.
Thanks in advance for your help!
Re: :READ? return in simultaneous voltage and current measurement
Post by Andrea C » January 16th, 2019, 5:37 am
You can send a format command for which data elements should be returned in response to the ":READ?" command.
Default is:
:FORM:ELEM VOLT,CURR,RES,TIME,STAT
If one of the parameters is not measured, e.g. resistance in your case, but is still in the format specifier for return, the instrument will report 9.91E37 to indicate not a number (NAN).
JeanSu
Post by JeanSu » January 29th, 2019, 1:01 am
perfect. thank you for your answer!
