Tektronix Technical Forums are maintained by community involvement. Feel free to post questions or respond to questions by other members. Should you require a time-sensitive answer, please contact your local Tektronix support center here.
Stop at compliance without sweep
Models 2400, 2401, 2410, 2420, 2425, 2430, 2440, 6430
Post Reply
noddle
Country: Netherlands
Post by noddle » April 1st, 2011, 12:03 am
Hi,
I use Keithley 2410 source-meter to do TDDB measurement. I want this source-meter stop at current compliance.
I have tried to put this code
:STAT:MEAS:ENAB 16384
*SRE 1
:ABOR
:STAT:MEAS?
*SRE 0
but it didnot work. Any help would be greatly appreciated.
Dale C
Keithley Applications
Country: United States
Contact:
Re: Stop at compliance without sweep
Post by Dale C » April 1st, 2011, 7:34 am
Are you doing a sweep?
If so then use the compliance abort command.
:SOURCE:SWEEP:CABORT (NEVER,EARLY or LATE).
If not using a sweep then you could also use the trip command.
:SENSE:CURRENT:PROTECTION:TRIPPED?
This would return a yes(1) or no(0) when addressed to talk.