Hi I Was exploring the following functions in 2600b manual:
smuX.measure.delay
smuX.source.delay
how can these be used to eliminate the delay() in the following code:
node[1].smua.source.func = node[1].smua.OUTPUT_DCAMPS
node[1].smua.source.output = node[1].smua.OUTPUT_ON
node[1].smua.source.rangei = 0.001000
node[1].smua.source.limiti = 0.0001000
node[1].smua.source.leveli = 0.0001000
delay (0.005000)
node[1].smua.measure.nplc =   0.1000
print(node[1].smua.measure.v())
node[1].smua.source.output = node[1].smua.OUTPUT_OFF
Vince W
Re: Exploring source delay and measure delay
Post by Vince W » November 23rd, 2016, 10:26 am
First, note that the default source and measure delay settings are smuX.DELAY_AUTO, which will choose a range-dependent delay based on the instrument's typical settling time characteristics. That said, if you want to explicitly control the source and measure delays, you can do so.
The source delay is the amount of time that the instrument waits after changing the output value for it to settle to the desired output level, before it continues to any other operation.
The measure delay is the amount of time that the instrument will wait for range change settling, etc. prior to starting a measurement.
So, these two settings are additive. In your code, you had a call to delay(.005) or 5 mS.
The following code splits that time between the source and measure delays.
Also, I would recommend setting all smu parameters prior to turning on the output, including starting the output level at 0.
node[1].smua.source.delay = 0.001000
node[1].smua.measure.delay = 0.004000