Check if SMU is Busy Sweeping
BiggyD
Post by BiggyD » June 20th, 2018, 2:47 pm
I'm using VB.NET and I'm send the following to my 2636B:
myInstr.IO.WriteString("smua.trigger.source.linearv(.01, 1, 11)")
myInstr.IO.WriteString("smua.trigger.source.action = 1")
myInstr.IO.WriteString("smua.trigger.count =20")
myInstr.IO.WriteString("smua.trigger.source.limitv = 2")
myInstr.IO.WriteString("smua.source.delay = .1")
myInstr.IO.WriteString("smua.trigger.endsweep.action = smua.SOURCE_IDLE")
myInstr.IO.WriteString("smua.trigger.arm.count = 2")
myInstr.IO.WriteString("smua.trigger.initiate()")
myInstr.IO.WriteString("waitcomplete()")
This works fine, but I need to wait until the sweep is done before moving on to the next command. How can I check to see if the sweep is done? :?:
brian.d.smith
Re: Check if SMU is Busy Sweeping
Post by brian.d.smith » July 5th, 2018, 7:01 am
From the Reference Manual page 3-30
To start a sweep, use the smuX.trigger.initiate() function. Sweeps are overlapped
operations, so you can use the waitcomplete() function as a way to suspend further operation
until the sweep is complete.
BiggyD
Post by BiggyD » July 5th, 2018, 9:48 am
Here's the solution:
Do
'retreive sweep status and place into buffer
myInstr.IO.WriteString("print(status.operation.sweeping.condition)")
'read buffer
statusInt1 = Val(myInstr.ReadString())
'retreive measurement status and place into buffer
myInstr.IO.WriteString("print(status.operation.measuring.condition)")
'read buffer
statusInt2 = Val(myInstr.ReadString())
'if either sweep or measurement is still in progress, continue loop
If statusInt1 = 0 And statusInt2 = 0 Then Exit Do
Loop