Post by Andrea C » May 9th, 2021, 9:32 am

As a sample exercise, this is a TSP script for using two SMU within a 2600A or 2600B for an LIV curve.

SMUA = applies voltages to the LED or Laser
SMUB = applies voltage bias to PD and measures current vs. time
LIV_on_4N35.PNG
LIV on opto coupler
LIV_on_4N35.PNG (12.26 KiB) Viewed 219 times
Using ASYNC trigger model, you can measure at up to 20K Samples/sec.

Here is some sample data at 10KHz measure rate on the 4N35:
2636B_LIV_samples_vs_time.PNG
sample outcome
2636B_LIV_samples_vs_time.PNG (85.74 KiB) Viewed 219 times
Here is the TSP script and pseudo code to make use of the script. Run it from Test Script Builder.
ASYNC_Sample_SrcList_4N35OptoCoupler.tsp
TSP Script File
(6.39 KiB) Downloaded 18 times