DAQ6510 Not Measuring Current
ulasus
Country: Turkey
Post by ulasus » August 3rd, 2021, 12:17 am
[*]Hello! I have a problem with the device Keithley DAQ6510. I am trying to obtain current from rear terminals, although it doesn't allow any terminal to measure current, it only has terminals for DC voltage. How can I modify it? Thank you.
Re: DAQ6510 Not Measuring Current
Post by Andrea C » August 10th, 2021, 10:30 am
Keep in mind that multimeters have a different input terminal for the Amps measure function vs for the voltage, ohms or temperature measure function.
When using the rear inputs and taking signals from one of the 77xx switch cards, only the cards that have dedicated amps channels will be able to connect your current into the Amps terminal. Those dedicated amps channels also provide pathway completion when signal is not being routed through the internal amp meter. If they did not have pathway completion, then there would be open circuit (no current) unless actively measuring the current.
If your 77xx card does not have any amps channels, consider using a small shunt resistor in series with your current. Connect a volts channel across the shunt resistor and get your current from V/R.
Users browsing this forum: Google [Bot] and 1 guest
