Tektronix Technical Forums are maintained by community involvement. Feel free to post questions or respond to questions by other members. Should you require a time-sensitive answer, please contact your local Tektronix support center here.

2182A readings have many shots and large fluctuations

Post Reply
hachiko
Posts: 18
Joined: May 15th, 2013, 6:02 am
Country: China

2182A readings have many shots and large fluctuations

Post by hachiko » August 24th, 2014, 1:27 am

I have 2182A input cable's 2 blades (black and red) shorted. With default settings (channel 1, slow rate, filter on), the readings I got were following, showing many shots and large fluctuations.
I don't know where these shots came from since the input is shorted with Keithley's cable. How can I eliminate these shots and reduce fluctuations?

[img]
reading1.PNG
reading1.PNG (14.01 KiB) Viewed 24084 times
[/img]
[img]
reading2.PNG
reading2.PNG (23.39 KiB) Viewed 24084 times
[/img]

Andrea C
Keithley Applications
Keithley Applications
Posts: 1427
Joined: October 15th, 2010, 10:35 am
Country: United States
Contact:

Re: 2182A readings have many shots and large fluctuations

Post by Andrea C » August 26th, 2014, 7:14 am

Do you still have the DeoxIT solution that ships with the 2182A?
Try cleaning the contacts per the procedure on page 1-13 of the 2182A User's Manual.

Are you using a clean copper clamp to create the short between the red and black spade lugs on the 2107 input cable?

hachiko
Posts: 18
Joined: May 15th, 2013, 6:02 am
Country: China

Re: 2182A readings have many shots and large fluctuations

Post by hachiko » August 27th, 2014, 9:05 am

What is the " DeoxIT solution"? I don't understand. I will try the cleaning with a sand paper.

Yes I am using a clean clamp for shorting the red and black spade lugs.

Andrea C
Keithley Applications
Keithley Applications
Posts: 1427
Joined: October 15th, 2010, 10:35 am
Country: United States
Contact:

Re: 2182A readings have many shots and large fluctuations

Post by Andrea C » August 27th, 2014, 4:29 pm

The DeoxIT is a contact cleaning solution that is provided with the instrument when new. It is usually packaged with the copper alligator clips. The 4-pin copper contacts in the LEMO connector on the 2182A can also oxidize. This cleaning fluid can be used to remove the oxide.
2182A_contact_cleaner.PNG
2182A_contact_cleaner.PNG (28.4 KiB) Viewed 24048 times

hachiko
Posts: 18
Joined: May 15th, 2013, 6:02 am
Country: China

Re: 2182A readings have many shots and large fluctuations

Post by hachiko » October 9th, 2014, 5:39 am

I cleaned the clips, but still got the shots time by time.

Dale C
Keithley Applications
Keithley Applications
Posts: 2849
Joined: June 10th, 2010, 6:22 am
Country: United States
Contact:

Re: 2182A readings have many shots and large fluctuations

Post by Dale C » October 9th, 2014, 12:51 pm

I ran the 2182A with a short on the input cable for 10,000 readings with your settings.
The result is as follows...
This is what you should be seeing.
No spikes.
Attachments
2812A test.png
2812A test.png (218.35 KiB) Viewed 23980 times

Post Reply

Return to “Low Voltage Measurements”

Who is online