Post by JCP2017 » May 30th, 2017, 3:29 pm

Hi,
When using a 2400 purely as a voltmeter (say, to measure hall voltage) can one reduce the leakage current into the DUT by shunting a resistor across In/Out Hi/Lo and then only connect Sense Hi and Lo to the sample while using remote sensing or would this cause unreliable readings?

Andrea C
Keithley Applications
Keithley Applications
Posts: 1244
Joined: October 15th, 2010, 10:35 am
Country: United States
Contact:

Re: Quasi 4-probe measurement

Post by Andrea C » June 11th, 2017, 5:18 am

Have you tried it?

Keep in mind that there is a max allowable delta v between the force and senses leads. Using them in this "uncoupled" way could result in some odd behaviors with certain complex loads.