I'm trying to query the 2651a through python / GPIB but am not successful.
I can, for example, set the voltage of the sourcemeter without problem. This works:
instrument.write("smua.source.levelv = 1")
However, when I try to query the voltage level with this line of code:
reading = instrument.ask("smua.measure.v()")
I get error code -420, Query UNTERMINATED
Is "smua.measure.v()" the right command to issue?
Re: Problem in using python to communicate with 2651a
Post by Dale C » October 9th, 2012, 6:04 am
Try the following commands...
smua.source.levelv=1
smua.source.output-smua.OUTPUT_ON
reading=smua.measure.v()
print("Voltage reading="..reading)
Kuser2651a
Post by Kuser2651a » October 9th, 2012, 12:42 pm
I've tried the following:
instrument.write("smua.source.output-smua.OUTPUT_ON")
reading = instrument.ask("reading=smua.measure.v()")
The first two commands work but not the last one. I still get error code -420, query UNTERMINATED.
Kenneth P
Post by Kenneth P » October 9th, 2012, 6:57 pm
You might want to try assigning the measurement string to a variable if your GPIB function call will allow and send as parameter:
WrtString='reading=smua.measure.v()'
prtStr = 'print(reading)'
gpib.send(instrAddr,WrtString)
gpib.send(instrAddr,prtStr)
ReadString = gpib.enter(instrAddr,256)
The GPIB .send and .enter functions are from a Keithley Python Library. In this case, the reading is assigned to a variable in the 2600A and then printed directly.
Kuser2651a
Post by Kuser2651a » October 10th, 2012, 8:28 am
Thanks. Where can I download the Keithley python module? And are there any example python scripts showing how to use it?
Kenneth P
Post by Kenneth P » October 12th, 2012, 6:58 am
If I am not mistaken, your function call:
instrument.ask() is a GPIB query. You used with a source measurement function call. The reason why your syntax does not work is because the 2600A does not issue data on the GPIB in this fashion. Dale and I have noted in our examples that you must use the print() function call to return any information onto the GPIB.
You must first assign the measurement to a local variable within the 2600A and then use the print() fucntion to return the reading to the controller.
instrument.write("reading=smua.measure.v()")
reading = instrument.ask("print(reading)")
Again, I caution you in using ( and ) within a function. I don't know whether your library will handle this properly. I believe it will but if not, assign the strings with ( and ) to a local variable and using the variable as a parameter in your function calls. I.E.
PrtStr = "reading=smua.measure.v()"
instrument.write(PrtStr)