Putting Keithley 2651A into TALK mode via IEEE
Model 2651A
NMaunsell
Country: Ireland
Post by NMaunsell » April 19th, 2013, 8:26 am
Does anyone know how to put a Keithley 2651A into TALK mode, cant get the measurement to output and observed that the 2651 stays in LISTEN mode. Can it be forced to TALK? Using IEEE and FLUKE METCAL
Re: Putting Keithley 2651A into TALK mode via IEEE
Post by Andrea C » April 19th, 2013, 3:00 pm
If I just use a utility to send GPIB commands one by one, then you can see the TALK annunciator come on when you address the instrument to talk (by executing the Enter or ibrd command). For example, send the *IDN? command; the LSTN comes on. Read from the instrument and then the TALK comes on.
You should be able to have similar LSTN and TALK if you send print(smua.measure.rangei) and then read back the current measure range that is in effect.
Return to “2651A High Current SourceMeter”