Hi.
Are you aware of any possibility to use the 2651A for characterizing semiconductors at a higher DC current, than +/- 20A, that is the maximum intrinsic capability of the SourceMeter? With other words, to use an external equipment, that will expand the current capability to over 20Amps on DC (something like a Power Amp) and should be an extension of the 8011 Test Fixture?
Thanks.
Jerry.
Jane S
Re: Using 2651A at Higher DC current
Post by Jane S » April 24th, 2013, 12:56 pm
Hi Jerry,
You can parallel 2x2651A to meet your requirements. In TSP Express (built in software), there's an option for you to combine the SMU.
It is a Download Java Runtime J2SE(TM) Environment.
http://www.keithley.com/support/keidoc_ ... e=Software
Jane
akinokoe
Post by akinokoe » October 10th, 2014, 12:15 pm
I have a question regarding the two meters in parallel. At the computer side, can the computer treat the two Keithley as one? E.g. Can I set the current at 40A, 10V at the computer side? We have some other instrument and control software with third party Keithley 2651A driver.
Best,
Aki
Abhay1
Country: India
Post by Abhay1 » October 13th, 2014, 12:05 am
Hi
When you are using two models of 2651 or any 26XX models. you need to communicate with only one (say MAsters). Other instrument will be controlled by Master Instrument using TSP link.
Post by Andrea C » October 13th, 2014, 5:06 am
Aki,
Do you need 40 Amp continuous or pulsed? I wanted to mention that a single 2651A can provide 20 Amps continuous or up to 50 Amps pulsed.
If you use a second 2651A to double the current, the second unit can be communicated with over the TSP-Link from the first unit. So no direct connection to the PC is needed. But you still must send the individual commands to each physical 2561A.