Using 2657A to do staircase voltage sweep for breakdown test, can the script exit early if the current limit is encountered?

Andrea C
Keithley Applications
Keithley Applications
Posts: 1612
Joined: October 15th, 2010, 10:35 am
Country: United States
Contact:

Re: Breakdown Test Sweep V, Exit on Current Limit

Post by Andrea C » May 26th, 2021, 11:04 am

Yes, that can be done.

Here is sweeping the voltage to 250V (1V steps) on a 10GΩ resistor and on a 200V Zener.
In the case of the zener, we exit the sweep early.

This same code will work with any of the 2600A or 2600B models (up to their max voltage capability).
2657A_Exit_on_Compliance.PNG
Exit on Compliance Voltage Sweep
2657A_Exit_on_Compliance.PNG (41.29 KiB) Viewed 839 times
Here is screen capture of calling the TSP code from Python language:
2657A_Exit_on_Compliance_Python.PNG
Calling TSP function from Python.
2657A_Exit_on_Compliance_Python.PNG (28.77 KiB) Viewed 839 times
The TSP source and a PDF describing the process are attached.
Attachments
VSweep_Exit_on_Compliance.pdf
Full code listing
(273.24 KiB) Downloaded 28 times
Abort_Sweep_on_Compliance_functions.tsp
TSP function definition code
(7.48 KiB) Downloaded 23 times