Doing C-V test by connecting Keithely 4200 with Keithley590
ItsDryHeat
Post by ItsDryHeat » August 4th, 2015, 6:25 pm
I am interested in doing C-V test using a 590 and change settings/receive outputs by a 4200. Connection from 590 to 4200 were verified and I was able to get data from 590 on 4200's computer using a plotter emulator.
The UTM was configured to get its User Library from KI590ulib, its User Module from CvSweep590.
Though 590 itself was able to measure the capacitance of my setup and display it on its front panel. The measurement reading from 4200 was always "-10022" under column "CvSweep590". No other readings were there. It seems that no sweep nor other actions were triggered by 4200. I changed input channel from CMTR1 to CMTR2,3,and4, it did not work. Can anyone help me figure out how to set 4200 correctly to conduct this cv test?
A screenshot of my settings and a screenshot of the output are attached to this post.
Thanks.
Reading from 4200.JPG (24.72 KiB) Viewed 16380 times
Vince W
Re: Doing C-V test by connecting Keithely 4200 with Keithley
Post by Vince W » August 5th, 2015, 12:44 pm
Hello, you may be my customer. Are you in AZ? We have a dry heat in CO as well, just not as HOT.
The value of the variable CvSweep590 is the return value of the UTM function. It is returning an error code, which are in the UTM description:
RETURN VALUES:
0 OK.
-10022 (KI590_NOT_IN_KCON) There is no CMTR defined in
your system's configuration.
The code is looking for "KI590" in the CMTR1 instrument ID's MODELNUM attribute. I'm checking this on my 4200. I'll get back here as soon as I test this out. I don't have a 590, but I don't think I'll need one.
Vince W
Post by Vince W » August 5th, 2015, 12:59 pm
So, this error did not occur on my system.
I think the problem is that you don't have CMTR1 in your 4200 system configuration (KCON).
Exit KITE, start KCON, select Tools->Add External Instrument->Capacitance Meter->Keithley 590 CV Analyzer
By default, this should create an instrument named CMTR1.
Make sure the GPIB address matches that of the 590 and there are no other instruments connected with that address.
Be sure to save the configuration using File->Save
ItsDryHeat
Post by ItsDryHeat » August 5th, 2015, 2:57 pm
Hi Vince,
Yep, I'm in AZ.
Thanks a lot, re-configuration solved the problem. Now everything is up and running.