Tektronix Technical Forums are maintained by community involvement. Feel free to post questions or respond to questions by other members. Should you require a time-sensitive answer, please contact your local Tektronix support center here.

Error: contact check fail

Post Reply
Andrea C
Keithley Applications
Keithley Applications
Posts: 1619
Joined: October 15th, 2010, 10:35 am
Country: United States
Contact:

Error: contact check fail

Post by Andrea C » December 31st, 2020, 7:28 am

When executing an ITM, if you experience an error for contact check fail, it is an indication that your remote sense lines are not connected. The error messages will include the Rhi and RLo measured values. These represent the measured ohms between the HI and Sense HI (Rhi) and between LO and Sense LO (RLo).

If you intend to use the remote sense lines (four wire connections), then the contact check fail issue should be resolved by correcting the connections.

If you do not need to use the remote sense lines, then to operate the SMU in two-wire mode, set the Sense Mode at device level to Local.
ACS_SenseMode_Device_Level.PNG
Device Level - Sense Mode Setting
ACS_SenseMode_Device_Level.PNG (25.76 KiB) Viewed 740 times
When it is set to default, it makes use of the settings in the Tools>Preferences (Misc tab).