TSP code for 10Amp Pulsed IV - Rds(ON) measurement
Model 2450, 2460
Locked
Post by Andrea C » March 26th, 2020, 2:10 pm
This posting is supplementary information and code listings pertaining to topics covered in a series of videos.
Video Links:
Part 1 Introduction to 10Amp Pulse for Rds(ON): https://www.youtube.com/watch?v=C_YqjNgO68Y
Part 2 Front Panel Setup and Compare to Scope: https://www.youtube.com/watch?v=ZIx0ERuzkew
Code run from Test Script Builder.
tested with:  KEITHLEY INSTRUMENTS,MODEL 2461,04110442,1.7.1e
Purpose:  apply current pulse from 2461, delay xx usec, digitize I and V for 100usec, pulse off.
          from buffer stats, get average v.
          compute average sourced current from sourcevalues buffer elements
          compute R from avg(V)/avg(I)
--Reset the instrument
reset()
eventlog.clear()
trigger.model.load("Empty")
     --setup digital output
     digio.line[1].mode = digio.MODE_TRIGGER_OUT
     trigger.digout[1].pulsewidth = 10e-6
     trigger.digout[1].stimulus   = trigger.EVENT_NOTIFY2
     -- end setup digital output
-- create source config list for the current pulse and voltage limits
ConfigListNameSrc = "myPulseSrc"
smu.source.configlist.create(ConfigListNameSrc)
--Set the source to output current.
smu.source.func = smu.FUNC_DC_CURRENT
smu.source.readback = smu.ON        -- dual digitizers, so measure the current too
smu.source.range = 10
smu.source.level = 0
smu.source.vlimit.level = 1
smu.source.pulse.level = 10
smu.source.pulse.vlimit.level = 2  -- voltage limit during the pulse
smu.source.configlist.store(ConfigListNameSrc)
-- create a measure config list
ConfigListNameMeas = "myPulseMeas"
smu.measure.configlist.create(ConfigListNameMeas)
--Set up the measure functions.
smu.digitize.func = smu.FUNC_DIGITIZE_VOLTAGE
smu.digitize.sense = smu.SENSE_4WIRE
smu.digitize.range = 2
smu.digitize.samplerate = 1e6
smu.digitize.aperture = smu.APERTURE_AUTO
defbuffer1.capacity = 100000  -- does not have to match the smaller acquistion we are doing
smu.measure.configlist.store(ConfigListNameMeas)
-- some vars
debug = 1
-- enter your desired measure duration;  number of samples (measCount) will be computed based on the sample rate in effect.
measDuration = 100e-6
measCount = measDuration * smu.digitize.samplerate
measDelay = 400e-6
-- pulse width will be measDelay + measDuration
if debug then print("Pulse Width: "..measDelay + measDuration) end
-- Timer for the measure delay
local N = 1
trigger.timer[N].enable = 0
trigger.timer[N].reset()
trigger.timer[N].clear()
trigger.timer[N].enable = 1
trigger.timer[N].delay = measDelay
trigger.timer[N].count = 1
trigger.timer[N].start.stimulus = trigger.EVENT_NOTIFY1
trigger.timer[N].start.generate = trigger.OFF  -- first tic only after timer elapses
-- program the trigger blocks
blockNumber = 1
trigger.model.setblock(blockNumber, trigger.BLOCK_BUFFER_CLEAR)
blockNumber = blockNumber + 1
trigger.model.setblock(blockNumber, trigger.BLOCK_SOURCE_OUTPUT, smu.ON)
trigger.model.setblock(blockNumber, trigger.BLOCK_CONFIG_RECALL, ConfigListNameSrc, 1, ConfigListNameMeas, 1)
trigger.model.setblock(blockNumber, trigger.BLOCK_NOTIFY, trigger.EVENT_NOTIFY1)  -- starts timer1
trigger.model.setblock(blockNumber, trigger.BLOCK_NOTIFY, trigger.EVENT_NOTIFY2)  -- strobe the digital output
trigger.model.setblock(blockNumber, trigger.BLOCK_SOURCE_PULSE_OUTPUT, smu.ON)  -- start 10Amp pulse
trigger.model.setblock(blockNumber, trigger.BLOCK_WAIT, trigger.EVENT_TIMER1)  -- wait for measure delay
trigger.model.setblock(blockNumber, trigger.BLOCK_MEASURE, defbuffer1, measCount)  -- 100 readings at 1MSample/sec
trigger.model.setblock(blockNumber, trigger.BLOCK_SOURCE_PULSE_OUTPUT, smu.OFF)  --stop 10Amp pulse
trigger.model.setblock(blockNumber, trigger.BLOCK_SOURCE_OUTPUT, smu.OFF)
---Initiate trigger model and wait until finished.
trigger.model.initiate()
waitcomplete()
statsVar = buffer.getstats(defbuffer1)
-- compute mean source value
-- we have source readback on
src_val_sum = 0
for i = 1, defbuffer1.n do
   src_val_sum = src_val_sum + defbuffer1.sourcevalues[i]
end  -- for loop
avg_src_val = src_val_sum / defbuffer1.n
display.clear()
display.changescreen(display.SCREEN_USER_SWIPE)
display.settext(display.TEXT1, "Rds(ON):  "..string.format("%.5f",statsVar.mean/avg_src_val))
display.settext(display.TEXT2, "Avg Src Current: "..string.format("%.5f",avg_src_val))
if debug then
  print("Rds(ON):  "..string.format("%.5f",statsVar.mean/avg_src_val))
  print("Avg V:  "..string.format("%.5f",statsVar.mean))
  print("Avg I:  "..string.format("%.5f",avg_src_val))
end
Re: TSP code for 10Amp Pulsed IV - Rds(ON) measurement
Post by Andrea C » May 11th, 2020, 2:44 pm
Video Link:
Part 3: https://www.youtube.com/watch?v=9n4mgW2Btig
Part 4: https://www.youtube.com/watch?v=mizxVYp-ezc
Code for Part 4:
*** Copyright 2020 Tektronix, Inc.                      ***
*** See www.tek.com/sample-license for licensing terms. ***
          Option to repeat N times (RepeatCount)
          2461 10Amp Pulsing has pulse width and duty cycle limits enforced by firmware:
             max PW using 10Amp range = 1msec
             max Duty Cycle for 10Amp range and compliance of 20V or less = 10%
             max Duty Cycel for 10Amp range and compliance >20V = 5%
             At lower currents/watts, different timing is possible.
function config_digital_line()
	     --setup digital output to strobe when NOTIFY2 event occurs
end   -- function
function createSourceConfigList(ConfigListNameSrc, irange, pulseLevel, pulseLimit)
		smu.source.range = irange
		smu.source.pulse.level = pulseLevel
		smu.source.pulse.vlimit.level = pulseLimit  -- voltage limit during the pulse
end  -- function
function appendSourceConfigList(ConfigListNameSrc, pulseLevel, pulseLimit)
end  -- function
function createMeasureConfigList(ConfigListNameMeas, vrange, sampleRate)
		smu.digitize.range = vrange
		smu.digitize.samplerate = sampleRate
end  -- function
function config_pulse_timers(pulsePeriodTime, NumPulses, measDelayTime, pulseWidthTime)
		-- ***********************   CONFIG TIMERS *************************
		-- for single pulse, this one not so important
		-- but for multiple pulses, this timer will
		-- manage the duty cycle
		pulsePeriod = trigger.timer[1]
		pulsePeriod.enable = 0
		pulsePeriod.reset()
		pulsePeriod.clear()
		pulsePeriod.enable = 1
		pulsePeriod.delay = pulsePeriodTime
		pulsePeriod.count = NumPulses   -- or how many entries in the source config list
		pulsePeriod.start.stimulus = trigger.EVENT_NOTIFY1  -- raised by our trigger blocks
		pulsePeriod.start.generate = trigger.OFF
		measDelay = trigger.timer[2]
		measDelay.enable = 0
		measDelay.reset()
		measDelay.clear()
		measDelay.enable = 1
		measDelay.delay = measDelayTime
		measDelay.count = 1
		measDelay.start.stimulus = trigger.EVENT_TIMER1  -- pulseperiod timer
		measDelay.start.generate = trigger.OFF  -- first tic only after timer elapses
		-- control when pulse output turns off
		pulseWidth = trigger.timer[3]
		pulseWidth.enable = 0
		pulseWidth.reset()
		pulseWidth.clear()
		pulseWidth.enable = 1
		pulseWidth.delay = pulseWidthTime
		pulseWidth.count = 1
		pulseWidth.start.stimulus = trigger.EVENT_TIMER1  -- pulseperiod timer
		pulseWidth.start.generate = trigger.OFF
		-- ***********************   END TIMERS *************************
end  -- function
function load_trigger_blocks(ConfigListNameSrc, ConfigListNameMeas, measCount, RepeatCount)
send_strobes = false
		blockNumber = 1
		trigger.model.setblock(blockNumber, trigger.BLOCK_NOTIFY, trigger.EVENT_NOTIFY1)  -- starts timer1, PW Timer
		branchBackToHere = blockNumber
		trigger.model.setblock(blockNumber, trigger.BLOCK_CONFIG_NEXT, ConfigListNameSrc, ConfigListNameMeas)
		trigger.model.setblock(blockNumber, trigger.BLOCK_WAIT, trigger.EVENT_TIMER1)  -- wait for pulse period timer
		if send_strobes then
		end
		trigger.model.setblock(blockNumber, trigger.BLOCK_WAIT, trigger.EVENT_TIMER2)  -- wait for measure delay
		end
		end
		trigger.model.setblock(blockNumber, trigger.BLOCK_WAIT, trigger.EVENT_TIMER3)  -- wait for pulse width timer
		end
		trigger.model.setblock(blockNumber, trigger.BLOCK_BRANCH_COUNTER, RepeatCount, branchBackToHere)
		-- assert SRQ
end  -- function
function handle_data(debug_flag)
		src_val_sum = 0
		end  -- for loop
		display.clear()
		rds = string.format("%.5f",statsVar.mean/avg_src_val)
		display.settext(display.TEXT1, "Rds(ON):  "..rds)
		if debug_flag then
		end
		return rds
end  -- function
function config_test(pw, dutycycle, DesiredMeasTime, smu_digitize_rate, numPulses, pulseHeight, expectedR, debug)
-- compute params
local v_limit_during_pulse = (expectedR * pulseHeight) * 1.2  -- give 20% headroom on expected v
local srcRange = pulseHeight     -- what source range to use 10A, 7A, 5A, 4A, 1A, 100mA, etc.
-- compute other vars based on the entered values
local measCount = DesiredMeasTime * smu_digitize_rate
local pulsePeriodDuration = pw / (dutycycle / 100)   -- 10% duty cycle for 10Amp range and 20 or less vlimit
local overhead = 20e-6  -- TODO - do I need this?
local measDelayDuration = pw - DesiredMeasTime - overhead
if debug then
        print("Meas Delay: "..measDelayDuration)
        print("Pulse Period: "..pulsePeriodDuration)
        print("Pulse Width: "..pw)
        print("number of source values: "..numPulses)
end
-- config smu
local ConfigListNameSrc = "myPulseSrc"
local ConfigListNameMeas = "myPulseMeas"
       -- config digital output to give us timing marks
	   -- create config lists
	   --function createSourceConfigList(ConfigListNameSrc, irange, pulseLevel, pulseLimit)
	   createSourceConfigList(ConfigListNameSrc, srcRange, pulseHeight, v_limit_during_pulse)
	   -- for future use in case the appendSourceConfigList is used for Pulsed Sweeps
	   -- how many entries are in our source config list = number of pulses
	   --RepeatCount = smu.source.configlist.size(ConfigListNameSrc)
	   --function createMeasureConfigList(ConfigListNameMeas, vrange, sampleRate)
	   createMeasureConfigList(ConfigListNameMeas, v_limit_during_pulse, smu_digitize_rate)
	   --function config_pulse_timers(pulsePeriodTime, NumPulses, measDelayTime, pulseWidthTime)
	   config_pulse_timers(pulsePeriodDuration, numPulses, measDelayDuration, pw)
	   -- load trigger model
	   --function load_trigger_blocks(ConfigListNameSrc, ConfigListNameMeas, measCount, RepeatCount)
	   load_trigger_blocks(ConfigListNameSrc, ConfigListNameMeas, measCount, numPulses)
end  -- function
function run_test(debug_flag)
	waitcomplete()  -- comment this out and wait SRQ if calling from Python
	return handle_data(debug_flag)  -- return the Rds(ON) value
end  -- function
-- *****************       Main Program Below    *******************
local first_run = true
if first_run then
		reset()
		eventlog.clear()
		status.clear()
end  -- if
-- *******  Enter some paramter values
local debug = true
--  *****************   Pulse Timing  ******************
--  enter your desired measure duration, pulse width and duty cycle
--  number of samples (measCount) will be computed based on the sample rate in effect.
local measDuration = 100e-6
local pulseWidthDuration = 540e-6
local DutyCyclePct = 10
local digitizeRate = 1e6
--  *********************  Pulse Analog info ***********
--  enter pulse analog info
--  ***************  Pulse Amplitude and Compliance Limits
local RepeatCount = 6  -- for signal averaging, Rds(ON) based on average of averages
local pulseHeight = 10  -- how many amps
local Max_Expected_R = 0.105   -- influence voltage measure range
-- call our config functions with the entered parameters
--function config_test(pw, dutycycle, DesiredMeasTime, smu_digitize_rate, numPulses, pulseHeight, expectedR, debug)
config_test(pulseWidthDuration, DutyCyclePct,
            measDuration, digitizeRate,
            RepeatCount, pulseHeight, Max_Expected_R,
            debug)
print( run_test(debug) )
Locked
Return to “2450/2460 Touchscreen SourceMeter”