Tektronix Technical Forums are maintained by community involvement. Feel free to post questions or respond to questions by other members. Should you require a time-sensitive answer, please contact your local Tektronix support center here.

KickStart version 2.0.5 not run project (DMM7510)

Use this forum for questions about Keithley KickStart software.
Post Reply
Ilya Grudovik
Posts: 25
Joined: October 30th, 2018, 1:54 pm
Country: Belarus

KickStart version 2.0.5 not run project (DMM7510)

Post by Ilya Grudovik » November 7th, 2018, 9:12 am

See attachment.
Attachments
Project.ZIP
(783.66 KiB) Downloaded 547 times
1.png
1.png (29.76 KiB) Viewed 8678 times

Ilya Grudovik
Posts: 25
Joined: October 30th, 2018, 1:54 pm
Country: Belarus

Re: KickStart version 2.0.5 not run project (DMM7510)

Post by Ilya Grudovik » November 7th, 2018, 10:06 am

DMM connect from LAN (DHCP)

Jennifer C
Keithley Applications
Keithley Applications
Posts: 49
Joined: September 17th, 2010, 7:55 am
Country: United States

Re: KickStart version 2.0.5 not run project (DMM7510)

Post by Jennifer C » November 14th, 2018, 2:56 pm

Hi Ilya,

It appears that the DMM7510 is taking the measurements, but is timing out, and is unable to respond in the time that KickStart expects a response. I am not certain if this is due to an issue with the instrument or the software. I am reporting this issue to the engineering team. I do note that reducing the sample count does appear to have some benefit when retrieving readings. Is it possible that you can reduce your sample count? For example, can you try 3 million readings? Can you share a bit more about your application? If you'd like to reach me directly, please send a private message through the forum or email [email protected].

Thanks,

Jennifer

Ilya Grudovik
Posts: 25
Joined: October 30th, 2018, 1:54 pm
Country: Belarus

Re: KickStart version 2.0.5 not run project (DMM7510)

Post by Ilya Grudovik » November 16th, 2018, 3:28 pm

Hi Jennifer,

I am very glad that you responded. I have already started to think that the forum is dead: (
Jennifer C wrote:
November 14th, 2018, 2:56 pm
Is it possible that you can reduce your sample count? For example, can you try 3 million readings?
Thank! It works like this :)
Can you share a bit more about your application?
I wanted to see the signal as an oscilloscope in real time. I thought the program and the device allow it. Just as I can watch it on the instrument itself. After all, 1MS / s * 18 bits = 18 Mbit, which is much less than 100 Mbit LAN.

However, I see that reading the buffer is very slow: (
Watch 18bit is much nicer on a large monitor and not on a small built-in screen.

Jennifer C
Keithley Applications
Keithley Applications
Posts: 49
Joined: September 17th, 2010, 7:55 am
Country: United States

Re: KickStart version 2.0.5 not run project (DMM7510)

Post by Jennifer C » November 19th, 2018, 7:34 am

Hi Ilya,

Yes, reading the buffer is slow. We are investigating what optimizations that we need to make on the instrument and in the software to improve this issue. I will also capture this as a defect issue. When we submit defects, we like to capture customer information. Would you please send your company or university name to [email protected]?

Thanks,

Jennifer

Ilya Grudovik
Posts: 25
Joined: October 30th, 2018, 1:54 pm
Country: Belarus

Re: KickStart version 2.0.5 not run project (DMM7510)

Post by Ilya Grudovik » November 20th, 2018, 2:58 am

Jennifer C wrote:
November 19th, 2018, 7:34 am
Yes, reading the buffer is slow. We are investigating what optimizations that we need to make on the instrument and in the software to improve this issue. I will also capture this as a defect issue.
Thank you :) Glad to hear me.

Can I call your specialist to the topic on the 7510 device itself. Otherwise I feel very lonely there with my questions :)
When we submit defects, we like to capture customer information. Would you please send your company or university name to [email protected]?
Send.

Post Reply

Return to “KickStart”

Who is online

Users browsing this forum: No registered users and 1 guest