2510 resistance measurements incorrect
Post Reply
cballester
Country: United States
Post by cballester » March 22nd, 2012, 8:37 am
The issue we are having is that sometimes the units report very low or even negative resistance. If you reboot the instrument, the problem goes away. It does not matter what device and what instrument we use, they all have this issue.
This happens on at least 7 instruments that we have, and they have firmware revisions ranging from A03 to A08.
Dale C
Keithley Applications
Contact:
Re: 2510 resistance measurements incorrect
Post by Dale C » March 22nd, 2012, 11:14 am
What are you trying to measure?
If you are measuring resistance of an RTD or Thermistor?
If so, what setpoint? 1kohms, 10kohms?, etc.
Post Reply
Return to “Other Keithley Instrument Products”