Current/Voltage Pulse Generator
mabattis
Country: Argentina
Post by mabattis » July 8th, 2013, 3:57 pm
I’m doing some research on CMOS ICs and need to test them for Latch-up compliance based on standard JEDEC No. 78D, so I’m looking for an instrument capable of generating current and voltage pulses with the following minimum specifications:
Minimum Current Range: 50mA to 250mA.
Minimum Voltage Range: 0 to 10V.
Rise/Fall Time: 5us < tr < 5ms.
Min. Pulse width = 2 x tr. (2 times rise/fall time)
Max. Pulse width = 1 s.
It needs to have the possibility to limit current/voltage.
It could be only one instrument or two independent instruments.
Thanks!
Regards,
Mat
IC Latch Up Test - JESD78D.pdf
JEDEC Standard
(207.35 KiB) Downloaded 1225 times
Re: Current/Voltage Pulse Generator
Post by Andrea C » July 11th, 2013, 5:45 am
Hi Mat,
Take a look at the 2600B Series of SourceMeter Instruments. These can be used in a multi-SMU configuration to accomplish the multi pin latch-up test.
You can control the pulse width but not the rise/fall time of the pulse generated by these instruments. But the pulses that they generate will comply with the not faster than 5usec and not slower than 5msec.
Andrea