Offset on 2520
Enrique
Country: Spain
Offset on 2520
Post by Enrique » February 11th, 2014, 12:23 am
I am testing the model 2520 with an 200 Ohm resistance.
I have used regular BNC cables with crocodile termination. I have connected the resistance to the Hi and Lo source of current pulses. I have used another BNC cable to measure the drop of voltage in the resistance in a regular Oscilloscope.
I always detect an offset of at least 0.5V. Do you know where is this offset coming from?
I have read in the manual that there is always some offset current but is negligible, However that offset seems too much for me.
Norbert W
Re: Offset on 2520
Post by Norbert W » February 11th, 2014, 1:18 am
the 500 mA range has an offset current spec of 0.45 mA. With a load of 200 Ohm this may result in an offset voltage up to 0.09 V.
In the 1A / 5A ranges this level could be 10 times as high - theoretically as these ranges will not fit your 200 Ohm load.
Not sure if you are using DC or short pulses for your measurements.
I would recommend to start with DC and check the levels. When additionally checking pulses I would recommend to use a low inductance resistor
(not wirewound) and the delivered low inductance ('blue') cables.
Depending on the results you may want to adjust cabling, ask for adjustment / calibration or worst case repair.
Enrique
Country: Spain
Re: Offset on 2520
Post by Enrique » February 11th, 2014, 2:42 am
Does Keithley have any tool capable of producing pulses of current in the submicrosecond scale and with an offset in the micro Ampere scale?
Thanks.
Enrique
Norbert W
Re: Offset on 2520
Post by Norbert W » February 13th, 2014, 6:18 am
Considering your pulse width and offset requirements and
keeping in mind you mentioned a resistive load of 200 Ohms
you might widen you selection while looking at a function generator like
the model 3390.
Applicability here depends on the known constant resistive load so a
50 Ohm voltage pulser could be of some interest.
If the above makes sense please have a look at:
http://www.keithley.com/support/keidoc_ ... cification