Two DMM7510s making simultaneous measurements?
DMM7510 7½-Digit Graphical Sampling Multimeter
Forum rules
<t></t>
aVictor
Post by aVictor » January 29th, 2016, 4:58 am
We have two DMM7510 units. We need to get the two units making simultaneous VDC measurements of two different circuits.
I tried Kickstart, but it seems to be limited to a single DMM7510.
Is there a way to synchronize two DMM7510 and get continuous VDC samples from both units?
There's no voltage trigger in this setup, but I would settle for stopping VDC measurements on both units simultaneously and writing down the VDC measurement value. If that's the route, how can I stop both units at the same time?
Joseph V
Re: Two DMM7510s making simultaneous measurements?
Post by Joseph V » February 19th, 2016, 11:34 am
If you want to synchronize two DMM7510's together the simplest way is to trigger the instruments using TSP-Link.
Attached are two .tsp files which set up each instrument for synchronizing measurements. Note this is a very basic application which has some variation between reading times however for the most part each instrument is acquiring a reading ~6 microseconds apart from one another.
UnitSyncNode2.tsp
(1.27 KiB) Downloaded 1585 times
UnitSyncNode1.tsp
(1.29 KiB) Downloaded 1583 times
Return to “DMM7510 Graphical Sampling DMM”