7510 digitizer
DMM7510 7½-Digit Graphical Sampling Multimeter
Forum rules
<t></t>
zvikae
Country: Israel
7510 digitizer
Post by zvikae » July 12th, 2016, 1:02 am
hi, when using the digitizer and "auto range" when does the range selection is done?
Norbert W
Re: 7510 digitizer
Post by Norbert W » July 12th, 2016, 4:34 am
here a excerpt from the reference manual regarding the dmm.digitize.range command:
When you assign a range value, the instrument selects a fixed range that is large enough to measure
the assigned value. The instrument selects the best range for measuring the maximum expected
value.
For example, for digitize current measurements, if you expect a reading of approximately 9 mA, set
the range to 9 mA to select the 10 mA range.
When you read this setting, you see the positive full-scale value of the measurement range that the
instrument is presently using.
Return to “DMM7510 Graphical Sampling DMM”
