When using the 7510 and KickStart sample count is limited to 100000, I need to measure at 1000000 sample per second for 5 seconds. can it be done? how?

brian.d.smith
Keithley Applications
Keithley Applications
Posts: 402
Joined: June 10th, 2013, 6:00 am
Country: United States

Re: 7510 and KickStart - Sample Count

Post by brian.d.smith » May 22nd, 2017, 6:57 am

zvikae wrote:
May 22nd, 2017, 12:55 am
When using the 7510 and KickStart sample count is limited to 100000, I need to measure at 1000000 sample per second for 5 seconds. can it be done? how?
Below is an answer from one of our factory apps engineers:

Not in KickStart. KickStart caps the sample count due to data handling issues. For now, the answer is no.

On the instrument you can. Change buffer size to 5,000,000. Then set count to 5,000,000. Push TRIGGER. Wait for it…. Takes a few seconds to tens of seconds.

You can do it in TSP Express…just allow sufficient time for the data to be ready…a few seconds longer than 5 seconds.