Hi,
I'm trying to use my 2182a as a voltmeter to do a zero current electrochemical measurement. I'm trying to measure the potential between a reference electrode (saturated calomel electrode) and my GaN sample submerged in solution. Is the Keithley 2182a capable of doing this experiment? If so, how should I connect it to my system?
I'm new to this field and really need some help... any reply is appreciated!
Thanks!
Albert
Tektronix Technical Forums are maintained by community involvement. Feel free to post questions or respond to questions by other members. Should you require a time-sensitive answer, please contact your local Tektronix support center here.
2182a performing zero current electrochemical measurement
-
- Keithley Applications
- Posts: 402
- Joined: June 10th, 2013, 6:00 am
- Country: United States
Re: 2182a performing zero current electrochemical measuremen
Page 2-3 of the 2182A Reference Manual has this paragraph
"Voltage measurements
The Model 2182 has two voltage measurement functions: DCV1 and DCV2. DCV1 is
available for input Channel 1, and DCV2 is available for Channel 2.
DCV1 (Channel 1) has five measurement ranges (10mV, 100mV, 1V, 10V, and 100V) and can
measure voltage from 1nV to 120V. DCV2 (Channel 2) has three measurement ranges (100mV,
1V, and 10V) and can measure voltage from 10nV to 12V. Accuracy for each channel is listed
in the specifications (Appendix A)."
You will connect channel 1 of the instrument to your device with the included input cable.
You select the voltage range using the up and down arrow keys on the front panel. Or you can set the instrument to auto range by pressing the auto key.
The instrument will come up in DCV1 mode at start up in factory default mode.
"Voltage measurements
The Model 2182 has two voltage measurement functions: DCV1 and DCV2. DCV1 is
available for input Channel 1, and DCV2 is available for Channel 2.
DCV1 (Channel 1) has five measurement ranges (10mV, 100mV, 1V, 10V, and 100V) and can
measure voltage from 1nV to 120V. DCV2 (Channel 2) has three measurement ranges (100mV,
1V, and 10V) and can measure voltage from 10nV to 12V. Accuracy for each channel is listed
in the specifications (Appendix A)."
You will connect channel 1 of the instrument to your device with the included input cable.
You select the voltage range using the up and down arrow keys on the front panel. Or you can set the instrument to auto range by pressing the auto key.
The instrument will come up in DCV1 mode at start up in factory default mode.