Hello,

I am using the VI 'K2182Read1.vi' attached to read the channel 1 of my 2182A Nanovoltmeter. My laptop is connected with the nanovoltmeter through a GPIB-USB cable, and I am using LabView 2013. Whenever I run the program to read a voltage signal, the 2182A generates a beep, and the front panel displays 'ERROR: -101' and then 'ERROR: -110', although the readings seem to be correct. I don't know what causes these errors. Please give me some advice if you know to fix my problem.

Thank you very much.
Attachments
K2182Read1.vi
(13.73 KiB) Downloaded 1541 times

Joseph V
Keithley Applications
Keithley Applications
Posts: 107
Joined: July 1st, 2014, 10:31 am
Country: United States

Re: 2182A error messages

Post by Joseph V » January 12th, 2015, 2:15 pm

Hello,

The -101 and -110 are both syntax errors so as such you may not have entered a parameter correctly or mistyped somewhere in your code, i.e using a "," instead of a "."
If you can save the vi for labview 2012 I would be able to look at the vi and see where the error is coming from.

-Joe

cspin2013
Posts: 4
Joined: January 12th, 2015, 12:56 pm
Country: United States

Re: 2182A error messages

Post by cspin2013 » January 14th, 2015, 1:53 pm

Hello Joe,

I just attached a screenshot the block diagram of my labview vi and a LV2012 version of my labivew vi.

Thanks a lot.
Attachments
block_diagram.png
block diagram of my vi
block_diagram.png (74.61 KiB) Viewed 21353 times
K2182Read1.vi
LabView2012 version
(14.48 KiB) Downloaded 1558 times

Joseph V
Keithley Applications
Keithley Applications
Posts: 107
Joined: July 1st, 2014, 10:31 am
Country: United States

Re: 2182A error messages

Post by Joseph V » January 15th, 2015, 6:47 am

Hello

On the third write block you have it as :SENS:VOLT:CHAN1:RANG:5e-3 where is should be :SENS:VOLT:CHAN1:RANG 5e-3 so that extra colon is what was giving you the error.
Attached is the fixed VI.

-Joe
Attachments
K2182Read1 (1).vi
(13.49 KiB) Downloaded 1572 times

cspin2013
Posts: 4
Joined: January 12th, 2015, 12:56 pm
Country: United States

Re: 2182A error messages

Post by cspin2013 » January 15th, 2015, 2:25 pm

Hello Joe,

Thanks for correcting my syntax error. Actually I just copied that command from the quick reference guide that came with the nano-voltmeter.
When I tried to set auto-range, both ':SENS:VOLT:CHAN1:RANG:AUTO ON' and ':SENS:VOLT:CHAN1:RANG AUTO ON' generated message 'Error:-101'.
Do you know how to fix this problem?