What is the maximum possible ohm reading that can be measured by the Keithley 6487?
Dale C
Keithley Applications
Contact:
Re: 6487 - - Ohm Measurement
Post by Dale C » May 29th, 2015, 9:01 am
In theory, the 7485 can source up to 500V.
and it can read down to 1e-14A.
So 500 V / 1e-14A = 50 e17 ohms.
But that is not practical.
So the offset current is about 0.4pF. So I would use that to find out the max resistance.
500V / 0.4pA = 1.25 e15 ohms.