Pulsed current measurements on 610C
2502, 6485, 6487 - Support
martintri
Country: Canada
Post by martintri » April 5th, 2016, 3:56 pm
I was hoping to use an old 610C to measure a pulsed current. The peak current will be on the order of microamps, but it will be only 5 us long and run at 1 kHz. From what I can gather from the manual I don't believe I can do it due to the long rise times (somewhere it was written 3 ms in "fast" mode for uA range). I actually measure something but I don't believe it is a correct average. My first question is, does it make sense to use this to measure such a pulsed and short signal? If not, can I use the current integration function to perform such a measurement? Thank you
PS I wasn't sure if this or the electrometer sub-forum was the appropriate place for this, I apologize if this is the wrong one.
brian.d.smith
Re: Pulsed current measurements on 610C
Post by brian.d.smith » April 7th, 2016, 4:54 am
A 5uS pulse with a 50% duty cycle would be a 100kHz signal. If you look at the specs for the 610C used as an amplifier, the frequency response is limited to 3kHz at any gain. At maximum gain, frequency response is only 100Hz.
I think that the 610C cannot be used as an amplifier at the speed you require. The 610C was designed as a DC or very low frequency AC instrument.
martintri
Country: Canada
Post by martintri » April 12th, 2016, 4:20 pm
Hi Brian,
I see, thank you. This is what I thought but I wasn't quite sure. Thanks for confirming.
cheers,
Martin