Getting value of I, V ans S for each reading in Diff. Cond.
6220, 6221 Current Sources
vaibhavinegi
Post by vaibhavinegi » July 25th, 2015, 4:56 am
Using 6221 and 2182A devices I am performing IV operation using Differential Conductance.
Using FORMat:ELEMents, TRACE:DATA command can returns string set of Time-stamp , Reading, Unit, Average Voltage, Source Value and Reading Number.
How can we get returns string as Source Value, Voltage(not Average Voltage) and SIEMens?
Vaibhavi Negi
Research Scholar
Delhi; India
Re: Getting value of I, V ans S for each reading in Diff. Co
Post by Andrea C » July 28th, 2015, 6:04 am
Check out Figure 5-14 in the Reference Manual for the 6221.
Each differential conductance reading is from a computation using the voltage response to three different current excitation levels.
The individual source values and voltage measurements that go into a single DCON measurement are not available when making use of the differential conductance mode of operation.
You can send the UNIT S command to get SIEMens.
vaibhavinegi
Post by vaibhavinegi » July 28th, 2015, 9:24 pm
Then how you get all readings in KI6220.exe software.
The data sheet generated by it have all value including Source value, Voltage , Average Voltage and Differential Conductance. :?: :?: :?: :?:
Vaibhavi Negi
Research Scholar
Delhi; India
Post by Andrea C » July 29th, 2015, 8:25 am
For the 6221, the format elements command is:
Perhaps you could invoke the NI I/O Trace capture tool and then run the KI6220 utility software to see what commands are being sent to the instrument from the software.
Return to “Current Sources”