Tektronix Technical Forums are maintained by community involvement. Feel free to post questions or respond to questions by other members. Should you require a time-sensitive answer, please contact your local Tektronix support center here.

rates changed to slow when read voltage through K6221 RS232

6220, 6221 Current Sources
Post Reply
hachiko
Posts: 18
Joined: May 15th, 2013, 6:02 am
Country: China

rates changed to slow when read voltage through K6221 RS232

Post by hachiko » July 10th, 2013, 9:09 pm

I set 2182A rates to Medium first, then I read a measurement through command "SYST:COMM:SER:SEND MEASURE:VOLT:DC?", the rates is automatically changed to slow, why?
2182A is connected to 6221 through RS232, 6221 is GPIB connected to PC.

And I found, It is very slow to get reading from 2182A through 6221 RS232, it takes about 2-3 seconds.

Dale C
Keithley Applications
Keithley Applications
Posts: 2849
Joined: June 10th, 2010, 6:22 am
Country: United States
Contact:

Re: rates changed to slow when read voltage through K6221 RS

Post by Dale C » July 11th, 2013, 6:58 am

The :MEASURE:VOLT:DC? query command resets the function of DCV to *RST conditions.
I would use the following code instead.

:SENSE:FUNC 'VOLT:DC'
:READ?

That should keep the DCV function in MEDium speed.

hachiko
Posts: 18
Joined: May 15th, 2013, 6:02 am
Country: China

Re: rates changed to slow when read voltage through K6221 RS

Post by hachiko » July 11th, 2013, 7:35 am

Thanks, I got it.

Post Reply

Return to “Current Sources”