hello, I have a Keithley 617 electrometer. According to the manual, if you use its internal voltage source, there is a maximum current limit of 2mA for the simultaneous current measure. My question is: if I don't care the accuracy of the current reading, can i still use the voltage source even if the current is overloading? Is this going to destroy the electrometer in the long run? Thanks.
Re: 2mA current limit
Post by Dale C » November 9th, 2011, 5:44 am
When the current is "overloading" the voltage source is no longer a constant voltage source.
It is now a source of current with a 4mA value.
the voltage source was not designed to operate in the "compliance" or "overload" mode.
Keithley's SMUs (Model 2400 and 2600A series) do operate in the compliance mode.