7065 hall effect card column 5 connection
7001, 7002 and Switch Cards
claudeahyi
Post by claudeahyi » January 25th, 2012, 3:04 pm
How to connect to column 5 ?
Re: 7065 hall effect card column 5 connection
Post by Dale C » January 26th, 2012, 5:42 am
The commands to close cross points are 1!2!5.
with the 1 being the slot.
the 2 being the row.
and 5 being the column.
So this command closes row 2 and column 5 in slot 1.
Look in the instruction manual on page 3-5 Table 3-3 for details.
Here is the link to the manual on the KI web site.
http://www.keithley.com/support/data?asset=1118
claudeahyi
Post by claudeahyi » January 26th, 2012, 8:43 am
Sorry, actually I was asking for how to physically connect to column 5. 4 triax connectors go to the rows, 2 triax go to column 3 and 4 etc. However in the user manual, the connection to column 5 is not clearly stated.
Post by Dale C » January 26th, 2012, 11:02 am
Column 5 is the terminal strip on the back of the card.
Look at the instruction manual page 2-4 for the terminal strip table.
Look at the instruction manual page 3-1 and figure 3-7 for details.