7065 verification

Post by ramgen » June 29th, 2010, 10:14 am

What is a simple method to verify the Model 7065 is operating properly?
Which cross points should be closed to just output a current from the current source and be measured by the -.
Just to verify things are working and configured properly.

Dale C
Keithley Applications
Keithley Applications
Posts: 2821
Joined: June 10th, 2010, 6:22 am
Country: United States
Contact:

Re: 7065 verification

Post by Dale C » June 29th, 2010, 10:41 am

To verify the Model 7065 is operating properly use the following procedure.
1. Connect the - to the ammeter port on the card. This is the female BNC connector.
2. Connect the current source to the current source female triax connector on the card. This is the left most triax connector when viewing the card from the rear.
3. Place the 7065 into the 7001 or 7002 scanner mainframe. Make sure the slot the 7065 is in is set for the 7065 on the 7001 or 7002. Select CARD, TYPE, SLOT, and input 7065.
4. Close cross points 1!1!1 and 1!4!1.
5. Configure the current source to output 1mA.
6. Configure the - to the 1mA range and disable Zero Check.
7. There should be 1mA on the display of the -.
This verifies the connections are correct and all three units are operating properly.

simonw
Posts: 1
Joined: May 7th, 2015, 1:32 pm
Country: Canada

Re: 7065 verification

Post by simonw » May 7th, 2015, 1:54 pm

Dale C wrote:To verify the Model 7065 is operating properly use the following procedure.
1. Connect the - to the ammeter port on the card. This is the female BNC connector.
2. Connect the current source to the current source female triax connector on the card. This is the left most triax connector when viewing the card from the rear.
3. Place the 7065 into the 7001 or 7002 scanner mainframe. Make sure the slot the 7065 is in is set for the 7065 on the 7001 or 7002. Select CARD, TYPE, SLOT, and input 7065.
4. Close cross points 1!1!1 and 1!4!1.
5. Configure the current source to output 1mA.
6. Configure the - to the 1mA range and disable Zero Check.
7. There should be 1mA on the display of the -
This verifies the connections are correct and all three units are operating properly.
I am having an issue with the high resistivity mode in my 7065. In low resistivity mode all cross points behave properly and switch correctly, however, in high resistivity mode I am seeing very strange results (unstable and much higher voltages). I am beginning to suspect the guard source followers are fried. Is there a simple way to check this and is it repairable? Thanks

Dale C
Keithley Applications
Keithley Applications
Posts: 2821
Joined: June 10th, 2010, 6:22 am
Country: United States
Contact:

Re: 7065 verification

Post by Dale C » May 8th, 2015, 8:08 am

Did you watch the video at this link?
Maybe that will help.
http://www.keithley.com/centralized_dis ... etid=55820

Can you configure a circuit with four resistors. In a square, each leg is a resistor.
The resistors are the same magnitude.
Make it high value resistors(>1M ohm).
Then watch the video. and see if that helps.

mfennell
Posts: 5
Joined: January 19th, 2011, 10:01 am
Country: United States

Re: 7065 verification

Post by mfennell » May 16th, 2017, 3:46 pm

We are bringing up a 7065 card we recently purchased and are seeing erratic behavior on the guard lines.

Can you post a fresh link to the video referred to above? The link you provided was:
http://www.keithley.com/centralized_dis ... etid=55820

The old link does not work any more.

Thanks!

Andrea C
Keithley Applications
Keithley Applications
Posts: 1104
Joined: October 15th, 2010, 10:35 am
Country: United States
Contact:

Re: 7065 verification

Post by Andrea C » May 28th, 2017, 5:40 am

The video is available here:
https://www.youtube.com/watch?v=aEj407pBX4E