TDS 3052C Measurement Indicators Fail to Turn On
jaime_g
Post by jaime_g » March 7th, 2017, 4:01 pm
Dear Forum,
I have been trying unsuccessfully to programmatically turn on the measurement indicators for a delay measurement.
Here is the commands send to the set up the scope:
HEAD OFF;
DIS:GRA GRID;
DIS:INTENSIT:WAVE 80;
HOR:RECORDL 10000;
ACQ:MODE SAM;
ACQ:STOPA SEQ;
MEASU:GAT SCRE;
SEL:CH1 ON;
CH1:COUP DC;
CH1:POS 2.5;
CH1:OFFS 0.0;
CH1:PRO 1;
CH1:SCA 1;
CH1:IMP MEG;
CH1:BAN FUL;
SEL:CH2 ON;
CH2:COUP DC;
CH2:POS -3.0;
CH2:OFFS 0.0;
CH2:PRO 1;
CH2:SCA 10;
CH2:IMP MEG;
CH2:BAN FUL;
HOR:SCA 400E-6;
HOR:DEL:TIM 800E-6;
TRIG:A:LEVEL 0.5;
TRIG:A:MODE NORM;
TRIG:A:TYPE EDGE;
TRIG:A:EDGE:SOURCE CH1;
TRIG:A:EDGE:SLOPE FALL;
TRIG:A:EDGE:COUP DC;
ACQ:STATE ON
After the signal is acquired the following commands are send to measure the delay:
HOR:SCA 400E-6;
HOR:DEL:TIM 800E-6;
MEASU:MEAS1:SOURCE1 CH1;
MEASU:MEAS1:SOURCE2 CH2;
MEASU:MEAS1:TYPE DELAY;
MEASU:MEAS1:DEL:DIRE FORW;
MEASU:MEAS1:DEL:EDGE1 FALL;
MEASU:MEAS1:DEL:EDGE2 FALL;
MEASU:REFL:METH PERC;
MEASU:REFL:PERC:MID 50;
MEASU:REFL:PERC:MID2 10;
MEASU:INDIC:STATE MEAS1;
MEASU:MEAS1:STATE ON
This is the scope trace
TDS3052C Trace_53259 Turn On Delay.jpg (1.96 MiB) Viewed 15973 times
Any help on how to programmatically turn on the measurement indicators would be appreciate it.
Regards,
Jaime Arevalo
Dave W
Re: TDS 3052C Measurement Indicators Fail to Turn On
Post by Dave W » March 8th, 2017, 1:38 pm
Hi Jamie,
The page in the programmer's manual for the MEASUrement:INDICators:STATE command states
There must be an active measurement before you can activate an indicator for a specified measurement.
In your command sequence you are trying to turn on the indicators before the command to turn on the measurement.
Reverse the order of those two commands and you should be good! :)
jaime_g
Post by jaime_g » March 9th, 2017, 9:56 am
Dave,
Thank you for your prompt response.
Reversing the order of the commands worked.
It is interesting to note that the original order of these commands works on the MDO3012 and MDO3014 but not on the TD3052C.
Regards,
Jaime
Dave W
Post by Dave W » March 9th, 2017, 4:41 pm
Hi Jamie,
You're welcome!
Yeah, I can't say why the TDS3000 requires the measurement be on first. It was developed a long time ago! You're likely not the first person to experience the issue though which I'm sure is why they made the order not matter on the newer scopes.