Posted: April 9th, 2019, 9:33 pm
by micjor375
I have TBS102B Oscilloscope

I'm able to data log upto having timescale has 25ms/div.After 25ms/div,the oscilloscope is waiting for trigger
(i tried all trigger - channel 1,ext. trigger,AC line)

I need to datalog having timescale at max ie 50s/div. or more than 1s/div.is it possible?

The oscilloscope stores 1 file/sec. is it possible to store in a single file for datalog duration (eg.30mins).whether is it possible to make oscilloscope store 1file/5 mins or any other.Because i need to datalog upto one hour.

Re: DATA Logging - TBS1202B

Posted: April 10th, 2019, 9:47 am
by David N
If you need to record data up to 1 hour, the TBS1202B is not the instrument to use. At the slowest time recording of 50 sec/div (slowest setting) and with 10 divisions, you can capture 500 seconds of data (50 x 10 =500) which is about 8 minutes of data. That is the maximum data length the TBS1202B will capture

Re: DATA Logging - TBS1202B

Posted: April 11th, 2019, 5:19 am
by micjor375
I'm using the data logging functionality of TBS1202B. With this only,i have limitation of having time scale upto 25ms/div.

Re: DATA Logging - TBS1202B

Posted: April 11th, 2019, 10:37 am
by David N
When using the data logger feature and you want to have the time/div slower than 25ms/div, set the trigger mode to "Normal".