Tektronix Technical Forums are maintained by community involvement. Feel free to post questions or respond to questions by other members. Should you require a time-sensitive answer, please contact your local Tektronix support center here.

Problem with deskew calibration fixture

Post Reply
guimimo
Posts: 4
Joined: April 11th, 2013, 5:18 am

Problem with deskew calibration fixture

Post by guimimo » March 13th, 2015, 4:56 am

Hello everyone,

I possess a Deskew Calibration Fixture (067-0405-00) and I do not know exactly how can I do to use it for calibrate my current (TCP0030) and voltage (TDP0500) probes. More, I would like to perform an automatic accurate deskew between these probes using this fixture with my scope DPO7054.
Maybe someone could help me to use my fixture? Is it possible to calibrate my current probe, and realize automatic deskew?

Thank your very much for your support.

Best regards,

Guillaume.

David N
Tektronix Applications
Tektronix Applications
Posts: 525
Joined: July 31st, 2008, 8:26 am

Re: Problem with deskew calibration fixture

Post by David N » April 3rd, 2015, 4:09 pm

Hello Guillaume,

The 067-0405-00 deskew fixture you have was not designed for use with the probes you mentioned. For the probes you mentioned, the deskew fixture is 067-1686-02. You can download the instruction manual for either fixture from Tek.com. Search for 071073005 for your fixture you have and search for 071187200 for the newer 067-1686-02 fixture.

Post Reply

Return to “Other or Discontinued Probes”

Who is online

Users browsing this forum: No registered users and 2 guests