TDS2024B vertical position isn't zero
john song
Country: Canada
Post by john song » November 7th, 2018, 1:03 pm
The TDS2024B has the error between vertical position and voltage level. It's almost same in the center of vertical. There are big different on the top or bottom vertical. Please see the enclosed -. The self-calibration is failed in step 7. The probe doesn't connect anything. All of four channel have same issue. How can I eliminate this error. Thanks.
Re: TDS2024B vertical position isn't zero
Post by David N » November 7th, 2018, 2:37 pm
When you run the "self calibration" has the scope been warmed up for 20 minutes or more? If yes, and it fails the self calibration, it most likely needs a complete instrument calibration to restore correct operation.
If you would like to have your unit calibrated, you can submit a request for service on Tek.com under the "Calibration & Services" tab and then click on "Request Repair Services" or call 1-800-835-9200 prompt 2.
john song
Country: Canada
Post by john song » November 7th, 2018, 2:46 pm
Hi David,
Thanks for your reply. The scope has already warm up for 20 minutes.
I found another issue is that the waveform will have a lot noise when the horizontal is less than 250us when I connect probe to "Probe Comp". Please see the enclosed -. Can this issue also be solved by complete instrument calibration? Thanks.
Post by David N » November 7th, 2018, 4:03 pm
On my instrument going from 250us/div to 100us/div does show a bit more in noise, but that looks to excessive. It may be a h/w issue in the instrument vertical input. A calibration may resolve it. If not, then it is a h/w related issue and would need repair.