AFG3000 MATLAB ICT Send Waveform 1
Carl M
Post by Carl M » November 13th, 2009, 5:46 pm
Use the table view of the samples on the instrument to verify the individual samples.
%% AFG3000 MATLAB ICT Send Waveform 1
% Date: 11-18-2008
% Send 14 bit waveform data to an AFG3000.
% PREREQUISITE EXAMPLES
% MATLAB ICT Control 1 (Hello World)
% COMPATIBILITY
% AFG3000, AFG3000B
% TESTED & DEVELOPED
% Microsoft Windows XP SP2
% TekVISA 3.3.2.7
% MATLAB Version 7.6.0.324 (R2008a)
% Instrument Control Toolbox Version 2.6
% GPIB: National Instruments PCMCIA-GPIB (ni488k.sys v2.6.0f0)
% AFG3252 FW 3.0.1
% Tektronix provides the following example "AS IS" without any guarantees
% or support.  This example is for instructional guidance only.
%% variables
visa_vendor = 'tek';
visa_address = 'GPIB0::3::INSTR';
% example waveform: sine wave using 12 sample points, integers between 0 to
% 16382 (range of AFG).  hex notation is easier to verify later
% (note: max value of afg3000 is '3FFE' not '3FFF')
wave = {'2000' '2FFF' '3BB6' '3FFE' '3BB6' '2FFF' '2000' '1000' '0449' ...
    '0000' '0449' '1000'};
wave = hex2dec(wave);
%% pre-processing
% encode variable 'wave' into binary waveform data for AFG.  This is
% the same as AWG5000B but marker bits are ignored.  see AWG5000B series
% programmer manual for bit definitions.
binblock = zeros(2 * length(wave), 1);
binblock(2:2:end) = bitand(wave, 255);
binblock(1:2:end) = bitshift(wave, -8);
binblock = binblock';
% build binary block header
bytes = num2str(length(binblock));
header = ['#' num2str(length(bytes)) bytes];
%% instrument communication
afg = visa(visa_vendor, visa_address);
fopen(afg);
fwrite(afg, '*rst;');
fwrite(afg, '*cls;');
% clear edit memory and set to 12 samples
fwrite(afg, ':trace:define ememory, 12;');
% send the data to edit memory
fwrite(afg, [':trace ememory,' header binblock ';'], 'uint8');
% set channel 1 to arb function found in edit memory
fwrite(afg, ':source1:function ememory;');
% turn on channel 1
fwrite(afg, ':output1 on;');
% gracefully disconnect
fclose(afg);
delete(afg);
clear afg;
Please discuss.
MrKel
Re: AFG3000 MATLAB ICT Send Waveform 1
Post by MrKel » July 8th, 2013, 2:57 pm
Hi, Im trying to make my own waveform and I just cant set it more than 245 samples... I have downloaded a manual with the commands of how to communicate with the AFG but I just cant make it work with more samples. Im using the code from here as a base.
Any help about it?
Carl M
Post by Carl M » July 19th, 2013, 10:57 am
MrKel wrote:Hi, Im trying to make my own waveform and I just cant set it more than 245 samples... I have downloaded a manual with the commands of how to communicate with the AFG but I just cant make it work with more samples. Im using the code from here as a base.
Any help about it?
sounds like a buffer problem. you'll have to increase the buffer of your visa object in matlab or use multiple writes with EOI off. I do this in my AWG example: http://www1.tek.com/forum/viewtopic.php?f=6&t=3217
Carl M
Post by Carl M » January 25th, 2017, 4:57 pm
rewrite to show normalization to dac values as discussed in topic: viewtopic.php?f=529&t=139184#p282112
date: 01-25-2017
demonstrate 'data:define' and 'data' remote programmatic commands for the
AFG3000 series and AFG2000 series.
The data following the binary block header is unsigned 16-bit integer
values, most significant byte first (big-endian). Bits 15 and 16 are
ignored as the AFG has 14-bits of dynamic range. The key concept is the
digital values are unit-less. Value 0 is the minimum output and value 16382
is the maximum output. Amplitude (and offset, and timing) are applied at
run-time based on the AFG settings.
MATLAB R2016a v9.0.0.341360 64-bit
Instrument Control Toolbox v3.9
NI-VISA v15.5
AFG2021 v1.1.9
%% constants
visa_vendor = 'ni';
visa_descriptor = 'tcpip::134.62.36.91::instr';
record_length = 12;
%% calculations
% example arbitrary function: 1-cycle sine
sample = 0:(record_length - 1);
vector = sin(2 * pi * sample / length(sample));
% normalize to dac values
m = 16382 / (max(vector) - min(vector));
b = -m * min(vector);
dac_values = uint16(m * vector + b);
dac_bytes = typecast(swapbytes(dac_values), 'uint8');
% binary-block header
bbh = sprintf('%u', length(dac_bytes));
bbh = sprintf('#%u%s', length(bbh), bbh);
afg = visa(visa_vendor, visa_descriptor);
fopen(afg);
fprintf('connected to...\n')
r = query(afg, '*idn?');
fprintf(r);
fprintf('\n')
fprintf('clearing event status register & flushing messages\n')
fwrite(afg, '*cls');
r = query(afg, '*esr?');
fprintf(' esr value: %s', r);
r = query(afg, 'system:error?');
fprintf(' msg: %s', r);
fprintf('\n')
cmd = sprintf('data:define EMEM,%u', record_length);
fprintf('writing: "%s"\n', cmd);
fwrite(afg, cmd);
fprintf('\n')
cmd = ['data EMEM,' bbh dac_bytes];
fprintf('writing bytes (hex)...\n');
for r = 1:length(cmd)
    fprintf('%02X ', cmd(r));
    if (mod(r,16) == 0)
        fprintf('\n')
    end
end
fprintf('\n')
fwrite(afg, cmd);
fprintf('\n')
fprintf('disconnecting...\n\n')
fclose(afg);
delete(afg);
fprintf('done\n')
MATLAB Command Window output
>> AFG3000_MATLAB_ICT_Send_Waveform_1
connected to...
TEKTRONIX,AFG2021,C010334,SCPI:99.0 FV:1.1.9
 esr value: 0
 msg: 0,"No error"
writing: "data:define EMEM,12"
 esr value: 0
 msg: 0,"No error"
64 61 74 61 20 45 4D 45 4D 2C 23 32 32 34 1F FF
2F FF 3B B5 3F FE 3B B5 2F FF 1F FF 10 00 04 49
00 00 04 49 0F FF
 esr value: 0
 msg: 0,"No error"
disconnecting...
done
NI IO Trace correlation to AFG2021 display