pyVisa too slow with RS232 TDK3034
Duzy_Piotr
Country: Poland
Post by Duzy_Piotr » January 28th, 2017, 9:21 am
import visa
rm = visa.ResourceManager()
list=rm.list_resources()
print(list) # list[0] is RS-232 reference
my_instrument = rm.open_resource(list[0])
print(my_instrument.query('*IDN?'))
#it works fine until this point, asking for one line of text only.
print(my_instrument.query('CURVE?'))
#After this oscilloscope is sending lots of data (1000 digits formatted as strings).
I get:
In python 3.4
"VI_ERROR_TMO (-1073807339): Timeout expired before operation completed."
In python 3.6
VI_ERROR_ASRL_OVERRUN "An overrun error occurred during transfer. A character was not read from the hardware before the next character arrived.
As far as I know, pyVisa is reading too slow when compare to speed of data sending. Same procedure is working when I am using pyVisa/GPIB, or pySerial/COM.
How can I make it faster, or what can I do to get that data by pyVisa/COM?
TDK 3034 Oscilloscope is using RTS/CTS synchronization method. How to set it on Visa? Would that help in this case?
Regards.
Dave W
Re: pyVisa too slow with RS232 TDK3034
Post by Dave W » February 1st, 2017, 3:57 pm
Hi Duzy_Piotr,
Using pyVisa with Serial COM requires a little bit of extra configuration code to make work properly. If you want to use flow control you will need to configure pyVisa in your code to use it. This can be done with the following code.
# Serial Port Settings
baudRate = 9600
flowCtrl = vc.VI_ASRL_FLOW_RTS_CTS
#flowCtrl = vc.VI_ASRL_FLOW_NONE
# Open session to instrument
lib = rm.visalib
scope = rm.open_resource(visaResourceAddr)
scope.timeout = 5000
# Some settings depend on which interface is being used
interface = lib.get_attribute(scope.session, vc.VI_ATTR_INTF_TYPE)[0]
if interface == vc.VI_INTF_ASRL:
    lib.set_attribute(scope.session, vc.VI_ATTR_ASRL_FLOW_CNTRL, flowCtrl)
    lib.set_attribute(scope.session, vc.VI_ATTR_ASRL_BAUD, baudRate)
Another issue you will run into is with the CURVE? query. PyVISA over COM relies on termination characters (i.e. '\r' or '\n') to detect the end of a message and terminate a read. If your waveform data contains the ASCII value for these termination characters (which is very likely) your read will terminate prematurely. You will need to disable termination characters during the read to get all the data back. You can do this with the following code.
# If using RS-232 then need to disable end on termination character while reading the image data
if lib.get_attribute(scope.session, vc.VI_ATTR_INTF_TYPE)[0] == vc.VI_INTF_ASRL:
    lib.set_attribute(scope.session, vc.VI_ATTR_ASRL_END_IN, 0)
# Readback the data
#< CURVE? Query code here>
# Re-enable end on termination character for RS-232
    lib.set_attribute(scope.session, vc.VI_ATTR_ASRL_END_IN, 1)
There is a PyVISA example on the forums that shows how to fetch screenshot data from the scope using a COM port. The same principles used in this example can be used for fetching waveform data.
Save Hardcopy to PC, PyVisa, TDS, TPS, TBS 1K-2K Series
Hope this helps!
