RSA306B EMCVu RBW issue
yo3fff
Country: Romania
Post by yo3fff » May 9th, 2020, 7:19 am
I need to perform an EMC measurement using RBW 1MHz but I can't set the RSA306B with RBW greater than 500kHz.
The frequency range is 80 - 110MHz.
I can see that the RSA 306B is capable of RBW 20Hz to 8MHz.
What is wrong?
Thank you.
BR,
Cristian
w2aew
Re: RSA306B EMCVu RBW issue
Post by w2aew » May 10th, 2020, 7:00 pm
Can you provide a .setup file? I don't see this problem (provided the Range limits are wide enough).
yo3fff
Country: Romania
Post by yo3fff » May 10th, 2020, 10:55 pm
Hi w2aew,
Please find attached the range limits csv file.
Thank you.
Cristian
BW500k.csv
(515 Bytes) Downloaded 5 times
w2aew
Post by w2aew » May 11th, 2020, 6:11 am
I was able to change it to 1MHz (see attached). There may be something else in your setup that is limiting this. Can you attach a EMCVu project setup file .emcsetup? (Setup Wizard > Save Project Setup)
BW500k1M.csv
(516 Bytes) Downloaded 5 times
yo3fff
Country: Romania
Post by yo3fff » May 11th, 2020, 11:00 pm
Hi w2aew,
Please find attached the setup file.
Thank you.
BR,
Cristian
xxx.zip
(64.13 KiB) Downloaded 5 times
w2aew
Post by w2aew » May 13th, 2020, 6:32 am
I found out from engineering that the CISPR 16.1 standard has a limit on Band C (30MHz to 1GHz) for the 6dB RBW range of 100k to 500k. Therefore, this limit was put into EMCVu.