RSA6100A Load Table 2, Table 3 correction file from PC
richjohnson
Post by richjohnson » January 11th, 2018, 8:38 am
This command only loads to Ext Loss Table 1. If the command is set to EDIT2 (Table 2) or EDIT3 (Table 3), the command loads Loss Table 1 as if EDIT1 is the command.
MMEMory:CALibration:LOAD:CORRection:EXTernal:EDIT<x> (No Query Form).
Not a show stopper but I want to confirm this operation.
Rich J
richjohnson
Re: RSA6100A Load Table 2, Table 3 correction file from PC
Post by richjohnson » January 11th, 2018, 9:15 am
OK, this command loads a Table 1, 2 or 3 csv correction file located on the RSA HD drive. So my title needs to be corrected to "correction file from RSA".
I also noticed, if I use the RSA GUI Correction File Editor and load a correction to Table 2 or Table 3, a subsequent MMEM:CAL:LOAD:CORR:EXT:EDIT2or3 "filename" SCPI command is sent it works. Otherwise loading to Table 2 or Table 3 only load Table 1 correction file.
Rich J
tekzilla
Post by tekzilla » January 15th, 2018, 7:13 am
Hi Rich,
Nice work interpolating the behavior of this command. It does indeed load a file from the RSA's hard drive, and does require a filename argument.
Can you clarify your final point about using the table loss editor in the RSA GUI? I'm unclear on what the observation was.
Return to “RSA5000/6000 Series”
