Hi everyone,

I set the V-Limit of Keithley 220 as 100V, but it cannot go past the current at which the voltage hits around 25V. Is there any other setup that might put a cap in voltage before the V-Limit we set on the front panel of 220? I'm running it via LabVIEW programs.

brian.d.smith
Keithley Applications
Keithley Applications
Posts: 402
Joined: June 10th, 2013, 6:00 am
Country: United States

Re: Keithley 220 Voltage limit?

Post by brian.d.smith » July 5th, 2018, 6:31 am

Maybe you are shorting and/or loading the GUARD? The datasheet says that total (output + guard) current cannot exceed 105mA.

Make sure that you use an adapter that drops guard and does not short guard to LO/SHIELD.

You must use an appropriate triax adapter at the rear of the instrument. Using a BNC adapter will likely ruin the triax jack on the rear of the instrument.