Question about uncertainties
Ohio_Guy
Post by Ohio_Guy » October 12th, 2020, 11:52 am
I have a simple question about uncertainty specifications for digital multimeters.
The specifications sheet for multimeters lists the absolute uncertainties for various functions. As an example, the specification for measuring DC voltage may look something like this:
Calibration interval: 1 year
Range: 200 mV DC
Uncertainty: ±(100 ppm of reading + 10 ppm of range)
So, for example, if the meter is displaying 150.0000 mV, the uncertainty would be ± 0.017 mV.
O.K., I understand that. But let’s say I only care about a change in voltage, and not the absolute voltage itself. As an example, let’s say I measure a change of 2.000 mV:
Starting voltage: 150.0000 mV
Ending Voltage: 152.0000 mV
Change in voltage: 2.000 mV
What is the uncertainty on the 2.000 mV? What equation or method would I use to calculate the uncertainty on the 2.000 mV change in voltage?
Re: Question about uncertainties
Post by Andrea C » October 21st, 2020, 6:13 am
From the specs, you can compute the error for each reading at the 150mV and 152mV levels.
Using your example, that is 17uV and 17.2uV
Do RMS on these to arrive at 24.4uV.
I believe RMS would be the error combining approach.
Worst case, suppose the one reading was at low limit of allowed error and second reading was at high limit of allowed error. That's effectively just adding the errors of each reading together to give you error of 34.6uV on their relationship. But statistically, you'd be having awful luck to have such a distribution.
Kiriakos
Country: Greece
Post by Kiriakos » November 11th, 2020, 11:35 pm
While the theoretical question it did receive an answer, there is also uncertainty specification for the DC voltage source.
Uncertainty specifications for digital multimeters this has importance only if you think to perform inspection of calibration.
There is a function at respectable multimeters, this named as Standard Deviation SD, this is a logging mode including Min / Max / SD.