2380-120 voltage controlled load disable

Series 2380 DC Electronic Loads
Post Reply
ggdecker
Posts: 4
Joined: July 21st, 2021, 11:06 am
Country: United States

2380-120 voltage controlled load disable

Post by ggdecker » July 26th, 2021, 6:10 am

I've been using battery mode to turn off the load at a specific voltage while discharging a battery. My current issue is that my battery capacity exceeds the time limit allowed by the battery test mode.

All I want the load to do is turn off if the voltage drops below a set value. Is there a way to do this from the front panel? The VON setting doesn't seem to do anything.

Andrea C
Keithley Applications
Keithley Applications
Posts: 1615
Joined: October 15th, 2010, 10:35 am
Country: United States
Contact:

Re: 2380-120 voltage controlled load disable

Post by Andrea C » July 27th, 2021, 11:26 am

I'd imagine the constant current mode of operation is required so that the battery is discharged at a known C rate.

If the required duration is longer than the 99999 seconds supported by the 2380 load, a SourceMeter such as model 2450 may serve you better. With the built-in scripting in the instrument, you could easily implement a loop until the measured V falls to XX level while sinking a prescribed amount of current to discharge a battery. It could even save the V vs. Time model to a thumb drive, if desired.